MS4525DO
PCB Mounted Digital Output Transducer
Combination Temperature and Pressure
C or SPI Protocol
Differential, Gage, Absolute, Compound, &
Vacuum
Temperature Compensated
3.3 or 5.0 Vdc Supply Voltage
Low Power Option Available (standby < 1uA)
The MS4525DO is a small, ceramic based, PCB mounted
pressure transducer from Measurement Specialties.
The
transducer is built using Measurement
Specialties’ proprietary
UltraStable™ process and the latest CMOS sensor conditioning
circuitry to create a low cost, high performance digital output
pressure (14bit) and temperature (11bit) transducer designed to
meet the strictest requirements from OEM customers.
The MS4525DO is fully calibrated and temperature compensated
with a total error band (TEB) of less than 1.0% over the
compensated pressure range. The sensor operates from single
supply of either 3.3 or 5.0Vdc and requires a single external
component for proper operation.
The rugged ceramic transducer is available in side port, top port,
and manifold mount and can measure absolute, gauge,
differential, vacuum or compound pressure from 1 to 150psi. The
1/8” barbed pressure ports mate securely with 3/32” ID tubing.
SENSOR SOLUTIONS
///MS4525DO
06/2017
Page 1
MS4525DO
PSI Pressure Ranges
PCB Mountable
Digital Output
Barbed Pressure Ports
APPLICATIONS
Blocked Filter Detection
Altitude and Airspeed Measurements
Medical Instruments
Fire Suppression System
Panel Meter
Air Movement/Environmental Controls
Pneumatic Controls
STANDARD RANGES (PSI)
Pressure
Absolute
Gage
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
Differential
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP ,MM
DS, MM
DS, MM
DS, MM
DS, MM
DS, MM
Compound
Vacuum
Option Availability
-F, -L, -M
-F, -L, -M
-F, -L, -M
-F, -L, -M
-F, -L, -M
-F, -L, -M
-F, -L, -M
-F, -L, -M
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP, DS
See Package Configurations: DS= Dual Side Port, SS= Single Side Port, TP= Top Port, MM= Manifold Mount
Only I
C Protocol is Available on “L” type Pin Styles; Reference Ordering Information for
Details
Pin Style “L” is only available SS and MM port types.
Pin Style “C” is only available SS,
TP and MM port types.
SENSOR SOLUTIONS
/// MS4525DO
06/2017
Page 2
MS4525DO
BLOCK DIAGRAM
ABSOLUTE MAXIMUM RATINGS
Supply Voltage
Output Current
Storage Temperature
Humidity
Overpressure
Burst Pressure
ESD
Solder Temperature
Conditions
= 25 °C
= 25°C
= 25°C
= 25 °C, both Ports
= 25 °C, Port 1
HBM
Min
2.7
+125
95
Not to Exceed 300
-4
+4
250°C, 5 sec max.
%RH
psi
psi
kV
Non Condensing
See Table 1
EN 61000-4-2
TABLE 1- BURST PRESSURE BY RANGE AND PACKAGE STYLE
Range
001
002
005
015
030
050
DS
90
SS, TP, MM
psi
psi
psi
psi
psi
psi
psi
psi
SENSOR SOLUTIONS
/// MS4525DO
06/2017
Page 3
MS4525DO
ENVIRONMENTAL SPECIFICATIONS
Mechanical Shock
Mechanical Vibration
Thermal Shock
Life
MTTF
Conditions
Mil Spec 202F, Method 213B, Condition C, 3 Drops
Mil Spec 202F, Method 214A, Condition 1E, 1Hr Each Axis
100 Cycles over Storage Temperature, 30 minute dwell
1 Million FS Cycles
>10Yrs, 70 ºC, 1.188 Million Pressure Cycles, 120%FS Pressure
PERFORMANCE SPECIFICATIONS
Supply Voltage
: 5.0V or 3.3 Vdc
Reference Temperature: 25°C (unless otherwise specified)
PARAMETERS
Pressure Accuracy
Total Error Band (TEB)
Temperature Accuracy
Supply Current
Compensated Temperature
Output Pressure Resolution
Output Temperature Resolution
Update Time
Start Time to Data Ready
Weight
Media
-25
MIN
-0.25
-1.0
+105
TYP
MAX
0.25
UNITS
%Span
%Span
ºC
ºC
ºC
bits
bits
mS
mS
grams
NOTES
Non-Corrosive Dry Gases Compatible with Ceramic, Silicon,
Borosilicate Glass, RTV, Gold, Aluminum and Epoxy. See
“Wetted Material by Port Designation” chart below.
1. Proper operation requires an external capacitor placed as shown in Connection Diagram. Output is not ratiometric to supply voltage.
2. The maximum deviation from a best fit straight line (BFSL) fitted to the output measured over the pressure range at 25C. Includes all
errors due to pressure non linearity, hysteresis, and non repeatability.
3. Total pressure error band includes all accuracy errors, thermal errors over the compensated temperature range and span and offset
calibration tolerances. For ideal sensor output with respect to input pressure and temperature, reference Transfer Function charts below.
TEB values are valid only at the calibrated supply voltage.
4. The deviation from a best fit straight line (BFSL) fitted to the output measured over the compensated temperature range.
5. For errors beyond the compensated temperature range, see Extended Temperature Multiplier chart below.
6. Start time to data ready is the time to get valid data after POR (power on reset). The time to get subsequent valid data is then specified
by the update time specification.
7. This product can be configured for custom OEM requirements, contact factory for lower power consumption or higher accuracy.
SENSOR SOLUTIONS
/// MS4525DO
06/2017
Page 4
MS4525DO
CONNECTION DIAGRAM
1.
Place 100nF capacitor between Supply and GND to within 2 cm of sensor.
PRESSURE AND TEMPERATURE TRANSFER FUNCTION
Gage, Differential and Compound Pressure Types
3FFF
3CCB
399A
7FF
733
666
Output( 2 HEX counts)
599
4CC
266
199
CC
0333
0000
pe
Ty
pe
Ty
Pmax
Pmin
Pressure
-50 °C
Tmin
-10 °C
85 °C
150 °C
Tmax
A Type:
Output( Decimal counts) =
80% *16383
__________
* (Pressure
applied
-Pmin) + 10%*16383
Pmax-Pmin
90%*16383
__________
Pmax-Pmin
* (Pressure
applied
-Pmin) + 5%*16383
Temperature
Output (Decimal Counts)
(Output °C - (-50°C)
Tmin
) * 2047
_________________________________
(150°C
Tmax
- (-50°C)
Tmin
B Type:
SENSOR SOLUTIONS
/// MS4525DO
06/2017
Page 5