includes common-mode rating of: 1500V rms between input
and output and input and power; 250V rms between power and
output – no return path is required between the power and
signal output commons.
Filtering and Protection
– An optimized five-pole Butterworth
filter (with a 4-Hz bandwidth) provides 116dB of normal-mode
rejection (noise on signal) at 60 Hz and 108 dB at 50 Hz. Output
noise is an exceptionally low 0.3 Mv p-p in a 100 kHz
bandwidth and 6 mV p-p in a 5 MHz bandwidth.
The 5B35 protects the computer side from damage due to field-
side over-voltage faults. The module withstands 240V rms at
the input terminals without damage, thereby shielding the
internal computer-side circuitry from field-side over-voltage
conditions. In addition, the 5B35 is mix-and-match and hot
swappable with all other 5B Series modules, so can be inserted
or removed from any socket in the same backplane without
disrupting system power.
A series output switch eliminates the need for external
multiplexing in many applications. The switch is turned on by
an active-low enable input. If the switch is to be on at all times,
the enable-input should be grounded to power common as it is
on the 5B01 and 5B08 backplanes.
Figure 2
Rev. 0 | Page 2 of 9
5B35
5B35 Models Available
Model
5B35-01
5B35-02
5B35-03
5B35-04
5B35-05
5B35-C-01
Input Range
Platinum,
α
=0.00385*
-100°C to +100°C (-148°F to +212°F)
0°C to +100°C (+32°F to +212°F)
0°C to +200°C (+32°F to +392°F)
0°C to +600°C (+32°F to +1112°F)
-100°C to +200°C (-148°F to +392°F)
Copper,
α
=0.004274
0°C to +120°C (10
@ 0°C) (+32°F to
+248°F)
0°C to +120°C (10
@ 25°C)(+32°F to
+248°F)
Nickel,
α
=0.00672
0°C to +300°C (+32°F to +572°F)
Output Range
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
Accuracy
±0.26°C
±0.13°C
±0.26°C
±0.78°C
±0.60°C
±0.23°C
5B35-C-02
5B35-N-01
5B35-Custom
5B35-C-Custom
5B35-N-Custom
0 V to +5 V
0 V to +5 V
±0.23°C
±0.40°C
* Custom Input/Output ranges are available including versions for the 100
Platinum RTD with
α
=0.003916. Refer to
configuration guide.
5B35 Specifications
(typical @ +25°C and V
s
= +5 V dc)
Input Ranges
Standard Ranges
Custom Ranges
Refer to Model Table
-200°C to +850°C (100
Platinum,
α
= 0.00385)
-80°C to +320°C (120
Nickel,
α
= 0.00672)
-100°C to +260°C (10
Copper,
α
= 0.004274)
0 V to +5 V
Accuracy
Initial @ +25°C
Conformity Error
Input Offset vs. Temperature
Output Offset vs. Temperature
Gain vs. Temperature
±0.05% Span
±0.01°C/°C
±20 µV/°C
±0.003% of Reading/°C
Input Resistance
Power On
Power Off
Overload
> 1000 MΩ
200 kΩ
200 kΩ
Noise
Input, 0.1 Hz to 10 Hz Bandwidth
0.2 µV rms
Rev. 0 | Page 3 of 9
Model 5B35
Output Range (R
> 50 kΩ)
5B35
Output, 100 kHz Bandwidth
Output, 1 MHz Bandwidth
Output, 5 MHz Bandwidth
Bandwidth, -3 dB
Output Rise Time, 10% to 90% Span
100 µV rms (0.3 mV peak-peak)
1.5 mV peak-peak
6 mV peak-peak
4 Hz
100 ms
Common-Mode Voltage (CMV)
Input-to-Output, Continuous
Input-to-Power, Continuous
Output-to-Power, Continuous
Transient
1500 V rms, maximum
1500 V rms, maximum
250 V rms, maximum
ANSI/IEEE C37.90.1-1989
Common Mode Rejection (CMR)
1 kΩ Source Imbalance, 50/60 Hz
Normal Mode Rejection (NMR)
190 dB (180 dB, Model 5B35-04)
116 dB @ 60 Hz; 108 dB @ 50 Hz
Sensor Excitation Current
Pt, 120
Ni
Cu
0.25 mA
1.0 mA
Lead Resistance Effect
Pt, 120
Ni
10Ω Cu
±0.00001°C/Ω
±0.0001°C/Ω
Input Protection
Continuous
Transient
Output Resistance
Voltage Output Protection
Output Current Limit
Output Selection Time
240 V rms, maximum
Continuous Short to Ground
±9 mA
6 µs to ±1 mV of V
out
@ C
load
= 0 to 2,000 pF
Output Enable Control
Max Logic "0"
Min Logic "1"
Max Logic "1"
Input Current "0", "1"
Power Supply Voltage
Power Supply Current
Power Supply Sensitivity, RTI
Mechanical Dimensions
+0.8 V
+2.4 V
+100 V
0.5 µA
+5 V dc ±5%
15 mA
±0.2°C/% of V
s
2.275" x 2.375" x 0.595"
(57.8 mm x 59.1 mm x 15.1 mm)
Environmental
Temperature Range
Rated Performance
-40°C to +85°C
Rev. 0 | Page 4 of 9
5B35
Operating
Storage
Relative Humidity
RFI Susceptibility
-40°C to +85°C
-40°C to +85°C
0 to 93% @ +40°C non-condensing
±0.5% Span error @ 400 MHz, 5 Watt, 3 ft
Includes the combined effects of repeatability, hysteresis, and conformity error. Loads heavier than 50 k
will degrade conformity and gain temperature coefficient.
The user's board layout must separate Power Common from Output Common. When the output switch is not used, connect the Enable Input to the Output Common. Output
Common is connected to power common when the 5B35 is installed on a 5B Series backplane.