首页 > 器件类别 > 传感器
ABPLANN016KGAA5
Board Mount Pressure Sensors
器件类别:传感器
厂商名称:Honeywell
厂商官网:http://www.ssec.honeywell.com/
ABPLANN016KGAA5 在线购买
供应商:
器件:ABPLANN016KGAA5
价格:-
最低购买:-
库存:点击查看
点击购买
Honeywell
Basic Board Mount Pressure Sensors
ABP Series—High Accuracy, Compensated/Amplified
60 mbar to 10 bar | 6 kPa to 1 MPa | 1 psi to 150 psi
Digital or Analog Output, Liquid Media Capable
32305128
Issue E
Datasheet
improving resolution and system accuracy. Optional internal
diagnostics validate that the sensor readings are correct.
• Flexible: Supply voltage range, variety of pressure units,
types, and ranges, output options, and wide operating
temperature range simplify use in the application.
• Versatile: Wet-media compatibility, sleep mode, and
temperature output options make the sensor a versatile
choice for Internet of Things applications.
• Honeywell Brand: Utilize proprietary Honeywell technology,
and are protected by multiple global patents.
DIFFERENTIATION
• Enhanced accuracy helps the design engineer fully
understand the error in measurement.
• Wide supply voltage range offers design flexibility.
• Power consumption when utilizing sleep mode option allows
for use in battery-powered applications.
The ABP Series are piezoresistive silicon pressure sensors
offering a ratiometric analog or digital output for reading
pressure over the specified full scale pressure span and
temperature range. They are calibrated and temperature
compensated for sensor offset, sensitivity, temperature
effects and accuracy errors (which include non-linearity,
repeatability and hysteresis) using an on-board Application
Specific Integrated Circuit (ASIC). Calibrated output
values for pressure are updated at approximately 1 kHz for
analog and 2 kHz for digital. All products are designed and
manufactured according to ISO 9001 standards.
• Dry gases option: The input port is limited to non-corrosive,
non-ionic media (e.g., dry air, gases) and should not be
exposed to condensation. The gases are limited to media
compatible with high temperature polyamide, silicone,
alumina ceramic, silicon, gold, and glass.
• Liquid media option: Includes an additional silicone-based
gel coating to protect the electronics under port P1, which
enables use with non-corrosive liquids (e.g. water and
saline) and in applications where condensation can occur.
Since port P2 is designed for use with non-corrosive liquids,
this option is often suitable for wet-wet differential sensing.
VALUE TO CUSTOMERS
• Simplifies design-in: Small size saves room on the PC board
(PCB), or simplifies design in smaller and lower power
devices. Meets Moisture Sensitivity Level 1 requirements,
which allows for unlimited shelf life when stored at
<30 ºC/85 %RH and, under most storage conditions, allows
for PCB soldering without any material concern about solder
joint quality due to aging of the sensor terminals, which
minimizes the concern about aging of the terminals prior
PCB assembly. Pressure choices allow engineers to select
range required for their application. Leadless SMT, SMT, and
DIP package options.
• Cost-effective: Small size helps engineers reduce design
and manufacturing costs while maintaining enhanced
performance and reliability of the systems they design.
• Accurate: Total Error Band (TEB) and wide pressure range
enable engineers to optimize system performance by
• Measure gage and differential pressures
• Total Error Band (see Figure 1): ±1.5 %FSS
• Liquid media option: Allows for wet/wet operation on dual
ported devices
• Industry-leading long-term stability: ±0.25 %FSS
• Industry-leading accuracy: ±0.25 %FSS BFSL
• Wide pressure range: 60 mbar to 10 bar | 6 kPa to 1 MPa |
1 psi to 150 psi
• As small as 8 mm x 7 mm
• High burst pressures (see Table 7)
• Calibrated over temperature range of 0°C to 50°C [32°F to 122°F]
• Operates from a single power supply of either 3.3 Vdc or 5.0 Vdc
• Output: Ratiometric analog or I
C- or SPI-compatible 12-bit
digital
• Power consumption: 2 uA typical when utilizing sleep mode option
• Meet IPC/JEDEC J-STD-020D.1 Moisture Sensitivity Level 1
requirements
• REACH and RoHS compliant
• Options: Internal diagnostic function, liquid media, sleep
mode, temperature output
POTENTIAL APPLICATIONS
Medical:
CPAP, blood analysis, blood pressure monitoring,
breast pumps, drug dosing, hospital beds, massage
machines, oxygen concentrators, patient monitoring, sleep
apnea equipment, urine analyzers, ventilators/portable
ventilators, and wound therapy
Industrial:
Air brakes, HVAC/transmitters, life sciences,
material handling, pneumatic control and regulation,
process gas monitoring, and valve positioning/positioners
Commercial:
Air beds, coffee makers, and washing
machines
PORTFOLIO
The Basic ABP Series is an addition to the following portfolio
of board mount pressure sensors: Basic NBP Series, Basic
TBP Series, TruStability™ HSC Series, TruStability™ SSC
Series, TruStability™ TSC Series, and TruStability™ NSC Series.
Sensing and Internet of Things
Basic Board Mount Pressure Sensors, ABP Series
Figure 1. Total Error Band
Total Error Band (TEB) is a single specification that includes
all possible sources of error. TEB should not be confused with
accuracy, which is actually a component of TEB. TEB is the worst
error that the sensor could experience. The TEB specification
on a datasheet may be confusing. Honeywell uses the TEB
specification in its datasheet because it is the most comprehensive
measurement of a sensor’s true accuracy. Honeywell also
provides the accuracy specification in order to provide a common
comparison with competitors’ literature that does not use the
TEB specification. Many competitors do not use TEB—they simply
specify the accuracy of their device. Their accuracy specification,
however, may exclude certain parameters. On their datasheet, the
errors are listed individually. When combined, the total error (or
what would be TEB) can be significant.
Table 1. Absolute Maximum Ratings
Characteristic
Supply voltage (V
supply
Voltage on any pin
Digital interface clock frequency:
SPI
ESD susceptibility (human body model)
Storage temperature
Min.
-0.3
-0.3
-40 [-40]
6.0
supply
+ 0.3
85 [185]
Vdc
kHz
kV
°C [°F]
All Possible Errors
Offset
Full Scale Span
Pressure Non-Linearity
Pressure Hysteresis
Pressure Non-Repeatibility
Thermal Effect on Offset
Thermal Effect on Span
Thermal Hysteresis
Accuracy
BFSL
Total
Error
Band
Soldering time and temperature:
lead solder temperature (DIP)
4 s max. at 250°C [482°F]
peak reflow temperature (Leadless SMT, SMT)
15 s max. at 250°C [482°F]
Absolute maximum ratings are the extreme limits the device will withstand without damage.
Table 2. Environmental Specifications
Characteristic
Humidity:
all external surfaces
internal surfaces of Liquid Media Option (T and V)
internal surfaces of Dry Gases Option (N and D)
Vibration
Shock
Life
Solder reflow
0 %RH to 95 %RH, non-condensing
0 %RH to 100 %RH, condensing
15 g, 10 Hz to 2 kHz
100 g, 6 ms duration
1 million pressure cycles minimum
J-STD-020-D.1 Moisture Sensitivity Level 1
(unlimited shelf life when stored at <30°C/85 %RH)
Life may vary depending on specific application in which the sensor is used.
Table 3. Wetted Materials
Pressure Port 1 (P1)
Dry Gas Option
Liquid Media Option
Ports and covers
Substrate
alumina ceramic
Adhesives
epoxy, silicone
epoxy, silicone gel
silicon, glass, solder, gold,
304 SST
Electronic components
aluminum
Contact Honeywell Customer Service for detailed material information.
Component
Table 4. Sensor Pressure Types
Pressure Type
Gage
Differential
Output is proportional to the difference between applied pressure and atmospheric (ambient) pressure.
Output is proportional to the difference between the pressures applied to each port (Port 1 – Port 2).
Pressure Port 2 (P2)
alumina ceramic
epoxy, silicone
silicon
Table 5. Operating Specifications
Characteristic
Supply voltage (V
supply
1, 2, 3
3.3 Vdc
5.0 Vdc
Supply current:
3.3 Vdc
5.0 Vdc
sleep mode option
Operating temperature range
Compensated temperature range
Temperature output option
Startup time (power up to data ready)
Response time
Clipping limit:
upper
lower
I C/SPI voltage level:
low
high
Pull up on SDA/MISO, SCL/SCLK, SS
Total Error Band
Accuracy
Long term stability (1000 hr, 25°C [77°F])
Output resolution
Analog
Min.
3.0
4.75
-40 [-40]
0 [-32]
Typ.
5.0
2.1
2.7
3.6
5.25
2.8
3.8
85 [185]
50 [122]
±1.5
±0.25
±0.25
Min.
3.0
4.75
-40 [-40]
0 [-32]
Digital
Typ.
5.0
3.1
3.7
3.6
5.25
3.9
4.6
85 [185]
50 [122]
±1.5
±0.25
±0.25
Vdc
uA
°C [°F]
°C [°F]
ms
ms
%Vsupply
%Vsupply
kOhm
%FSS
%FSS BFSL
%FSS
0.03
%FSS
bits
Sensors are either 3.3 Vdc or 5.0 Vdc based on the catalog listing selected.
Ratiometricity of the sensor (the ability of the device output to scale to the supply voltage) is achieved within the specified
operating voltage.
The sensor is not reverse polarity protected. Incorrect application of supply voltage or ground to the wrong pin may cause
electrical failure.
Operating temperature range: The temperature range over which the sensor will produce an output proportional to pressure.
Compensated temperature range: The temperature range over which the sensor will produce an output proportional to
pressure within the specified performance limits.
Temperature output option: Typical temperature output error over the compensated temperature range of 0°C to 50°C.
Operation in Sleep Mode may affect temperature output error depending on duty cycle. Refer to Figure 2 for temperature
output error over the operating temperature range.
Total Error Band: The maximum deviation from the ideal transfer function over the entire compensated temperature and
pressure range. Includes all errors due to offset, full scale span, pressure non-linearity, pressure hysteresis, repeatability,
thermal effect on offset, thermal effect on span, and thermal hysteresis.
Full Scale Span (FSS): The algebraic difference between the output signal measured at the maximum (Pmax.) and minimum
(Pmin.) limits of the pressure range. (See Figure 3.)
Accuracy: The maximum deviation in output from a Best Fit Straight Line (BFSL) fitted to the output measured over the
pressure range at 25°C [77°F]. Includes all errors due to pressure non-linearity, pressure hysteresis, and non-repeatability.
Table 6. Sensor Output at Significant Percentages (Digital Versions Only)
% Output
Digital Counts
decimal
8192
14746
16383
hex
0x0000
0x0666
0x2000
0x399A
0x3FFF
Figure 2. Temperature Output Option Temperature Error (See Table 5.)
Error (°C)
-2.5
-40
Temperature (°C)
Figure 3. Transfer Function Limits
Analog Versions
Compensated Pressure Range
Output (%Vsupply)
Ideal
1.5% Total Error Band
min.
Output (V) =
Digital Versions
Pressure (example unit)
max.
0.8 x Vsupply
max.
– P
min.
x (Pressure
applied
– P
min.
) + 0.10 x Vsupply
Output (% of 2
counts)
Ideal
min.
max.
Output (% of 2
counts) =
80%
x (Pressure
applied
– P
min.
) + 10%
max.
– P
min.
Transfer Function “A” is shown. See Figure 4 for other available transfer functions.
Digital or Analog Output, Compensated/Amplified
Figure 4. Nomenclature and Order Guide
For example,
ABPDNNN150PGAA3
defines an ABP Series Amplified Basic Pressure Sensor, DIP package, NN pressure port, dry
gases only, no diagnostics,150 psi gage pressure range, analog output type, 10% to 90% of Vsupply (analog), transfer function,
no temperature output, no sleep mode, 3.3 Vdc supply voltage.
ABP
Product Series
ABP
Amplified Basic
NN
150PG
Transfer Function
Supply Voltage
3.3 Vdc
5.0 Vdc
Package
DIP (Dual Inline Pin)
SMT (Surface Mount Technology)
Leadless SMT
Vsupply (analog), 2
10% to 90% of output, no sleep modecounts (digital)
no temperature
10% to 90% of 2
temperature output counts (digital only) enabled
enabled, sleep mode
10% to 90% of 2 counts (digital only)
Pressure Port
DIP
SMT
Leadless SMT
no temperature output, sleep mode enabled
NN
No port
NN
No port
NN
No port
10% to 90% of 2
enabled, no sleep mode
Output Type
AN
Single axial
barbed port
AN
Single axial
barbed port
AN
Single axial
barbed port
Analog
SPI
C, Address 0x08
C, Address 0x18
C, Address 0x28
C, Address 0x38
C, Address 0x48
C, Address 0x58
C, Address 0x68
C, Address 0x78
Pressure Range
1, 2
60 mbar to 10 bar
6 kPa to 1 MPa
Differential
1 psi to 150 psi
Differential
LN
Single axial
barbless port
LN
Single axial
barbless port
LN
Single axial
barbless port
Differential
JN
Single radial
barbless port
JN
Single radial
barbless port
060MD
100MD
160MD
250MD
400MD
600MD
001BD
1.6BD
2.5BD
004BD
±60 mbar
±100 mbar
±160 mbar
±250 mbar
±400 mbar
±600 mbar
±1 bar
±1.6 bar
±2.5 bar
±4 bar
Gage
006KD
010KD
016KD
025KD
040KD
060KD
100KD
160KD
250KD
400KD
±6 kPa
±10 kPa
±16 kPa
±25 kPa
±40 kPa
±60 kPa
±100 kPa
±160 kPa
±250 kPa
±400 kPa
Gage
001PD
005PD
015PD
030PD
060PD
±1 psi
±5 psi
±15 psi
±30 psi
±60 psi
Gage
JJ
Dual radial
barbless ports,
same side
JJ
Dual radial
barbless ports,
same side
RN
Single radial
barbed port
RN
Single radial
barbed port
060MG
100MG
160MG
250MG
400MG
600MG
001BG
1.6BG
2.5BG
004BG
006BG
010BG
0 mbar to 60 mbar
0 mbar to 100 mbar
0 mbar to 160 mbar
0 mbar to 250 mbar
0 bar to 400 mbar
0 bar to 600 mbar
0 bar to 1 bar
0 bar to 1.6 bar
0 bar to 2.5 bar
0 bar to 4 bar
0 bar to 6 bar
0 bar to 10 bar
006KG
010KG
016KG
025KG
040KG
060KG
100KG
160KG
250KG
400KG
600KG
001GG
0 kPa to 6 kPa
0 kPa to 10 kPa
0 kPa to 16 kPa
0 kPa to 25 kPa
0 kPa to 40 kPa
0 kPa to 60 kPa
0 kPa to 100 kPa
0 kPa to 160 kPa
0 kPa to 250 kPa
0 kPa to 400 kPa
0 kPa to 600 kPa
0 kPa to 1 MPa
001PG
005PG
015PG
030PG
060PG
100PG
150PG
0 psi to 1 psi
0 psi to 5 psi
0 psi to 15 psi
0 psi to 30 psi
0 psi to 60 psi
0 psi to 100 psi
0 psi to 150 psi
Option
RR
Dual radial
barbed ports,
same side
RR
Dual radial
barbed ports,
same side
Dry gases only, no diagnostics
Dry gases only, diagnostics on
Liquid media, no diagnostics
Liquid media, diagnostics on
Custom pressure ranges are available. Contact Honeywell Customer Service for more
information.
See the explanation of sensor pressure types in Table 4.
The transfer function limits define the output of the sensor at a given pressure input.