MCP3550/1/3
Low-Power, Single-Channel 22-Bit Delta-Sigma ADCs
Features:
• 22-Bit ADC in Small 8-pin MSOP Package with
Automatic Internal Offset and Gain Calibration
• Low-Output Noise of 2.5 µV
RMS
with Effective
Resolution of 21.9 Bits (MCP3550/1)
• 3 µV Typical Offset Error
• 2 ppm Typical Full Scale Error
• 6 ppm Maximum INL Error
• Total Unadjusted Error Less Than 10 ppm
• No Digital Filter Settling Time, Single-Command
Conversions through 3-wire SPI Interface
• Ultra-Low Conversion Current (MCP3550/1):
- 100 µA Typical (V
DD
= 2.7V)
- 120 µA Typical (V
DD
= 5.0V)
• Differential Input with V
SS
to V
DD
Common Mode
Range
• 2.7V to 5.5V Single-Supply Operation
• Extended Temperature Range:
- -40°C to +125°C
Description:
The Microchip Technology Inc. MCP3550/1/3 devices
are 2.7V to 5.5V low-power, 22-bit Delta-Sigma
Analog-to-Digital Converters
ADCs).
The devices offer
output noise as low as 2.5 µV
RMS
, with a total
unadjusted error of 10 ppm. The family exhibits 6 ppm
Integral Non-Linearity (INL) error, 3 µV offset error and
less than 2 ppm full scale error. The MCP3550/1/3
devices provide high accuracy and low noise
performance
for
applications
where
sensor
measurements (such as pressure, temperature and
humidity) are performed. With the internal oscillator
and high oversampling rate, minimal external
components
are
required
for
high-accuracy
applications.
This product line has fully differential analog inputs,
making it compatible with a wide variety of sensor,
industrial control or process control applications.
The MCP3550/1/3 devices operate from -40°C to
+125°C and are available in the space-saving 8-pin
MSOP and SOIC packages.
Applications:
Weigh Scales
Direct Temperature Measurement
6-digit DVMs
Instrumentation
Data Acquisition
Strain Gauge Measurement
Package Types
MCP3550/1/3
MSOP, SOIC
REF
IN
+ 2
IN
– 3
SS
SS
DD
8 V
DD
7 CS
6 SDO/RDY
5 SCK
Block Diagram
REF
Serial Interface
IN
Internal
Oscillator
2005-2014 Microchip Technology Inc.
SINC
IN
3rd-Order
DS ADC
Modulator
w/ Internal
Calibration
SCK
SDO
RDY
CS
DD
POR
DS20001950F-page 1
MCP3550/1/3
NOTES:
DS20001950F-page 2
MCP3550/1/3
ELECTRICAL
CHARACTERISTICS
† Notice:
Stresses above those listed under "Absolute
Maximum Ratings" may cause permanent damage to the
device. This is a stress rating only and functional operation of
the device at those or any other conditions above those indi-
cated in the operation listings of this specification is not
implied. Exposure to maximum rating conditions for extended
periods may affect device reliability.
Absolute Maximum Ratings †
DD
...................................................................................7.0V
All inputs and outputs w.r.t V
SS
.... .......... -0.3V to V
DD
+ 0.3V
Difference Input Voltage ....................................... |V
DD
- V
SS
Output Short Circuit Current ................................ Continuous
Current at Input Pins ....................................................±2 mA
Current at Output and Supply Pins ............................±10 mA
Storage Temperature ....................................-65°C to +150°C
Ambient temp. with power applied ................-55°C to +125°C
ESD protection on all pins (HBM, MM)
kV,
400V
Maximum Junction Temperature (T
) . .........................+150°C
DC CHARACTERISTICS
Electrical Specifications:
Unless otherwise indicated, all parameters apply at -40°C
+85°C, V
DD
= 2.7V or 5.0V.
REF
= 2.5V. V
IN
+ = V
IN
- = V
CM
= V
REF
/2. All ppm units use 2*V
REF
as full scale range. Unless otherwise noted, specification
applies to entire MCP3550/1/3 family.
Parameters
Noise Performance (MCP3550/1)
No Missing Codes
Output Noise
Effective Resolution
Noise Performance (MCP3553)
No Missing Codes
Output Noise
Conversion Times
MCP3550-50
MCP3550-60
MCP3551
MCP3553
Accuracy
Offset Error
INL
OS
-12
Positive Full-Scale Error
Negative Full-Scale Error
Offset Drift
Positive/Negative Full-Scale Error
Drift
Note 1:
2:
3:
4:
5:
6:
FS,P
FS,N
-10
-10
±2
±3
±4
±6
±2
±2
0.040
0.028
+10
+10
ppm
µV
µV
µV
ppm
ppm
ppm/°C
ppm/°C
= +25°C only (Note
2)
= +25°C
= +85°C
= +125°C
= +25°C only
= +25°C only
CONV
CONV
CONV
CONV
-2.0%
-2.0%
-2.0%
-2.0%
73.1
16.67
+2.0%
+2.0%
+2.0%
+2.0%
ms
ms
ms
ms
NMC
e
ER
bits
µV
RMS
bits RMS
REF
= 5V
At DC (Note
5)
NMC
e
ER
22
21.9
bits
µV
RMS
bits RMS
REF
= 5V
At DC (Note
5)
Sym.
Min.
Typ.
Units
Conditions
This parameter is established by characterization and not 100% tested.
INL is the difference between the endpoint’s line and the measured code at the center of the quantization band.
This current is due to the leakage current and the current due to the offset voltage between V
IN
+ and V
IN
Input impedance is inversely proportional to clock frequency; typical values are for the MCP3550/1 device. V
REF
= 5V.
Characterized by design, but not tested.
Rejection performance depends on internal oscillator accuracy; see
Section 4.0 “Device Overview”
for more informa-
tion on oscillator and digital filter design. MCP3550/1 device rejection specifications characterized from 49 to 61 Hz.
DS20001950F-page 3
MCP3550/1/3
DC CHARACTERISTICS (CONTINUED)
+85°C, V
DD
= 2.7V or 5.0V.
REF
= 2.5V. V
IN
+ = V
IN
- = V
CM
= V
REF
REF
Parameters
Rejection Performance
(1,6)
Common Mode DC Rejection
Power Supply DC Rejection
Common Mode 50/60 Hz Rejection
Power Supply 50/60 Hz Rejection
CMRR
PSRR
PSRR
-135
-115
-135
-85
-120
dB
dB
dB
dB
dB
CM
varies from 0V to V
DD
MCP3551
only, V
DD
varies from
4.5V to 5.5V
MCP3550-50
or
MCP3550-60
only
at 50 or 60 Hz respectively, V
DD
varies from 4.5V to 5.5V
MCP3551
only,
0 < V
CM
< V
DD
-V
REF
< V
IN
= (V
IN
+ -V
IN
-) < +V
REF
MCP3550-50
or
MCP3550-60
only
0 < V
CM
< V
DD
-V
REF
< V
IN
= (V
IN
+ -V
IN
-) < +V
REF
CM
range from 0 to V
DD
Sym.
Min.
Typ.
Units
Conditions
Normal Mode 50 and 60 Hz
Rejection
Normal Mode 50 or 60 Hz
Rejection
NMRR
-85
dB
NMRR
-120
dB
Analog Inputs
Differential Input Range
Absolute/Common Mode Voltages
Analog Input Sampling Capacitor
Differential Input Impedance
Shutdown Mode Leakage Current
Reference Input
Voltage Range
Reference Input Sampling
Capacitor
Reference Input Impedance
Shutdown Mode Reference
Leakage Current
Power Requirements
Power Supply Voltage Range
MCP3550-50, MCP3551
Supply
Current
MCP3550-60, MCP3553
Supply
Current
Supply Current, Sleep Mode
Supply Current, Shutdown Mode
Serial Interface
Voltage Input High (CS, SCK)
Voltage Input Low (CS, SCK)
Voltage Output High (SDO/RDY)
Note 1:
2:
3:
4:
5:
6:
IH
IL
OH
0.7 V
DD
DD
- 0.5
0.4
OH
= 1 mA, V
DD
= 5.0V
DD
DD
DD
DDSL
DDS
2.7
170
185
µA
µA
µA
µA
µA
µA
CS = SCK = V
DD
DD
= 5V
DD
= 2.7V
DD
= 5V
DD
= 2.7V
0.1
DD
pF
MΩ
nA
Note 5
Note 4
IN
+ = V
IN
- = V
SS
; CS = V
DD
IN+
IN-
-V
REF
SS
- 0.3
+V
REF
DD
+ 0.3
pF
MΩ
nA
IN
+ = V
IN
- = V
DD
; CS = V
DD
(Note
3)
Note 5
IN
+ and V
IN
REF
= 5V.
for more informa-
DS20001950F-page 4
MCP3550/1/3
+85°C, V
DD
= 2.7V or 5.0V.
REF
= 2.5V. V
IN
+ = V
IN
- = V
CM
= V
REF
REF
Parameters
Voltage Output Low (SDO/RDY)
Input leakage Current
(CS, SCK)
Internal Pin Capacitance
(CS, SCK, SDO/RDY)
Note 1:
2:
3:
4:
5:
6:
Sym.
OL
LI
INT
Min.
-1
Typ.
0.4
Units
µA
pF
Note 1
Conditions
OH
= -1 mA, V
DD
= 5.0V
IN
+ and V
IN
REF
= 5V.
for more informa-
TEMPERATURE CHARACTERISTICS
Parameters
Temperature Ranges
Specified Temperature Range
Thermal Package Resistances
Thermal Resistance, 8L-MSOP
Thermal Resistance, 8L-SOIC
JA
JA
211
149.5
°C/W
°C/W
-40
+85
+125
Sym.
Min.
Typ.
Units
Conditions
SERIAL TIMINGS
+85°C,
DD
= 3.3V or 5.0V, SDO load = 50 pF.
Parameters
CLK Frequency
CLK High
CLK Low
CLK fall to output data valid
CS low to indicate RDY state
CS minimum low time
RDY flag setup time
CS rise to output disable
CS disable time
Power-up to CS LOW
CS High to Shutdown Mode
Sym.
f
SCK
HI
LO
DO
RDY
CSL
SU
DIS
CSD
PUCSL
CSHSD
Min.
90
90
Typ.
Units
MHz
ns
ns
ns
ns
µs
ns
ns
ns
µs
µs
Conditions
DS20001950F-page 5