首页 > 器件类别 >
NIPXI-6052E
Full-Featured E Series Multifunction DAQ 12 or 16-Bit, up to 1.25 MS/s, up to 64 Analog Inputs
厂商名称:NI(恩艾)
厂商官网:https://www.ni.com
E Series – Full-Featured
• 16 or 64 analog inputs at up to
1.25 MS/s, 12 or 16-bit resolution
• 2 analog outputs at up to 1 MS/s,
• 8 digital I/O lines (TTL/CMOS);
two 24-bit counter/timers
• Analog and digital triggering
• 14 or 15 analog input signal ranges
• NI-DAQ driver simplifies
configuration and measurements
Operating Systems
• Windows 2000/NT/XP
• Real-time performance with
LabVIEW (page 134)
• Others such as Linux and Mac OS X
(page 187)
Recommended Software
• LabVIEW
• LabWindows/CVI
• Measurement Studio
• VI Logger
Families
• NI 6071E
• NI 6070E
• NI 6062E
• NI 6052E
• NI 6040E
• NI 6033E
• NI 6032E
• NI 6031E
• NI 6030E
• NI 6020E (only digital triggering)
Other Compatible Software
• Visual Basic, C/C++, and C#
Driver Software (included)
• NI-DAQ 7
Calibration Certificate Included
See page 21.
Family
NI 6071E
NI 6070E
NI 6062E
NI 6052E
NI 6040E
NI 6033E
NI 6032E
NI 6031E
NI 6030E
NI 6020E
Bus
PCI, PXI
PCI, PXI, FireWire
PCMCIA
PCI, PXI, FireWire
PCI, PXI
PCI
PCI
PCI, PXI
PCI, PXI
NI USB
Analog
Inputs
64 SE/32 DI
16 SE/8 DI
16 SE/8 DI
16 SE/8 DI
16 SE/8 DI
64 SE/32 DI
16 SE/8 DI
64 SE/32 DI
16 SE/8 DI
16 SE/8 DI
Input
Resolution
12 bits
12 bits
12 bits
16 bits
12 bits
16 bits
16 bits
16 bits
16 bits
12 bits
Sampling Rate
1.25 MS/s
1.25 MS/s
500 kS/s
333 kS/s
500 kS/s
100 kS/s
100 kS/s
100 kS/s
100 kS/s
100 kS/s
Input
Range
±0.05 to ±10 V
±0.05 to ±10 V
±0.05 to ±10 V
±0.05 to ±10 V
±0.05 to ±10 V
±0.1 to ±10 V
±0.1 to ±10 V
±0.1 to ±10 V
±0.1 to ±10 V
±0.05 to ±10 V
Analog
Outputs
Output
Resolution
12 bits
12 bits
12 bits
16 bits
12 bits
16 bits
16 bits
12 bits
Output
Rate
1 MS/s
1 MS/s
850 kS/s
333 kS/s
1 MS/s
100 kS/s
100 kS/s
20 S/s
Output
Range
±10 V
±10 V
±10 V
±10 V
±10 V
±10 V
±10 V
±10 V
Digital I/O
Counter/Timers
2, 24-bit
2, 24-bit
2, 24-bit
2, 24-bit
2, 24-bit
2, 24-bit
2, 24-bit
2, 24-bit
2, 24-bit
2, 24-bit
Triggers
Analog, digital
Analog, digital
Analog, digital
Analog, digital
Analog, digital
Analog, digital
Analog, digital
Analog, digital
Analog, digital
Digital
Data Acquisition and
Signal Conditioning
Table 1. NI Full-Featured E Series Model Guide (See page 228 for detailed specifications.)
Overview and Applications
NI Full-Featured E Series devices are the fastest and the most accurate
multiplexed data acquisition devices available. They are ideal for
applications ranging from continuous high-speed data logging to
control applications to high voltage signal or sensor measurements
when used with NI signal conditioning. Synchronize the operations
of multiple devices using the RTSI bus or PXI trigger bus and easily
integrate other hardware such as motion control and machine vision
to create an entire measurement and control system.
Visit
ni.com/oem
for information on our
quantity discounts for OEM customers.
Highly Accurate Hardware Design
NI Full-Featured E Series DAQ devices includethe following features
and technologies:
Temperature Drift Protection Circuitry
– Designed with components
that minimize the effect of temperature changes on measurements to
less than 0.0006% of reading per °C.
Resolution-Improvement Technologies
– Carefully designed noise
floor maximizes resolution.
Onboard Self-Calibration
– Precise voltage reference included
for calibration and measurement accuracy. Self-calibration is
completely software controlled, with no potentiometers to adjust.
NI DAQ-STC
– Timing and control ASIC designed to provide more
flexibility, lower power consumption, and a higher immunity to noise
and jitter than off-the-shelf counter/timer chips.
National Instruments • Tel: (800) 433-3488 • Fax: (512) 683-9300 • info@ni.com • ni.com
199
NI MITE
– ASIC designed to optimize data transfer for multiple
simultaneous operations using bus mastering with three scatter-
gather DMA channels for maximum performance of concurrent
I/O operations.
NI PGIA
– Measurement and instrument class amplifier that
guarantees settling times at all gains. Typical commercial off-the-
shelf amplifier components do not meet the settling time
requirements for high-gain measurement applications.
PFI Lines
– Eight programmable function input (PFI) lines that
can be used for software-controlled routing of interboard and
intraboard digital and timing signals.
RTSI or PXI Trigger Bus
– Used to share timing and control signals
between devices and synchronize operations.
RSE Mode
– In addition to differential and nonreferenced single-
ended modes, NI full-featured E Series devices offer referenced
single-ended (RSE) mode for use with floating signal sources in
applications with channel counts higher than eight.
High-Performance Driver Engine
– Software-timed single-point
input (typically used in control loops) with NI-DAQ achieves
rates of up to 50 kHz. NI-DAQ also delivers maximum I/O
system throughput with a multithreaded driver.
Test Panels
– With NI-DAQ, you can test all of your device
functionality before you begin development.
Scaled Channels
– Easily scale your voltage data into the proper
engineering units using the NI-DAQ Measurement Ready virtual
channels by choosing from a list of common sensors and signals or
creating your own custom scale.
LabVIEW Integration
– All NI-DAQ functions use the waveform data
type, which carries acquired data and timing information directly
into more than 400 LabVIEW built-in analysis routines for display of
results in engineering units on a graph.
Worldwide Support and Services
NI provides you with a wealth of resources to help you get your
application up and running more quickly, including:
Technical Support
– Purchase of NI hardware or software gives you
access to application engineers all over the world as well as Web
resources with more than 3,000 measurement examples and more
than 9,000 KnowledgeBase entries. –
ni.com/support
NI Factory Installation Services (FIS)
– Software and hardware
installed in PXI and PXI/SCXI systems, tested and ready to
use –
ni.com/advisor
Calibration
– Includes NIST-traceable basic calibration certificates,
services for ANSI/NCSL-Z540 and periodic calibration –
ni.com/calibration
Signal Conditioning
Onboard Temperature Sensor
– Included for monitoring the
operating temperature of the device to ensure that it is operating
within the specified range.
Analog and Digital Triggering
– Only full-featured E Series devices
provide the ability to set a trigger based on the level of an analog
signal, in addition to the ability to trigger off an edge of a
digital signal.
More Input Ranges
– Up to 15 input ranges for optimal resolution,
even for signals smaller than 50 mV.
High-Performance, Easy-to-Use Driver Software
NI-DAQ is the robust driver software that makes it easy to access the
functionality of your data acquisition hardware, whether you are a
beginning or advanced user. Helpful features include:
Automatic Code Generation
– DAQ Assistant is an interactive guide
that steps you through configuring, testing, and programming
measurement tasks, and generating the necessary code automatically
for use in LabVIEW, LabWindows/CVI, or Measurement Studio.
Cleaner Code Development
– Basic and advanced software functions
have been combined into one easy-to-use yet powerful set to help
you build cleaner code and move from basic to advanced
applications without replacing functions.
Extended Warranty
– Meet project life-cycle requirements
and maintain optimal performance in a cost-effective way –
ni.com/services
Data Acquisition Training
– Instructor-led courses –
ni.com/training
Professional Services
– Feasibility, consulting, and integration
through our Alliance Partners –
ni.com/alliance
For more information on NI services and support,
please visit
ni.com/services
For information on device support in
NI-DAQ 7, visit
ni.com/dataacquisition
Models
NI 6030E, NI 6031E,
NI 6032E, NI 6033E
Measurement Sensitivity* (mV)
0.0023
Nominal Range (V)
Positive FS
Negative FS
1.147
2.077
-2.5
0.836
0.422
0.215
0.25
-0.25
0.2
-0.2
0.102
0.1
-0.1
0.061
0.05
-0.05
0.2
0.1
NI 6052E
NI 6070E, NI 6071E
0.0025
0.009
NI 6040E
0.008
Low-Cost E Series
NI 6034E, NI 6036E
NI 6023E, NI 6024E,
NI 6025E
0.0036
0.008
Basic
PCI-6013, PCI-6014
0.004
Absolute Accuracy (mV)
15.373
7.56
5.697
1.79
3.859
1.556
0.789
0.399
0.405
0.176
0.100
0.0611
7.269
5.645
2.271
1.146
0.583
0.247
0.135
4.747
0.876
1.190
0.479
0.243
0.137
0.064
0.035
1.232
2.119
0.850
0.428
0.242
0.111
0.059
14.369
5.193
3.605
1.452
0.735
0.379
0.163
0.091
6.765
5.391
2.167
1.092
0.558
0.235
0.127
16.504
5.263
0.846
0.106
8.984
2.003
0.471
0.069
Note: Accuracies are valid for measurements following an internal calibration. Measurement accuracies are listed for operational temperatures within ±1 °C of internal calibration temperature and ±10 °C of external or factory-calibration temperature. One-year
calibration interval recommended. The Absolute Accuracy at Full Scale calculations were performed for a maximum range input voltage (for example, 10 V for the ±10 V range) after one year, assuming 100 pt averaging of data.*Smallest detectable voltage change in
the input signal at the smallest input range.
Table 2. E Series Analog Input Absolute Accuracy Specifications
NI 6052E
NI 6070E, NI 6071E
Low-Cost E Series
PCI-6036E
PCI-6024E, NI 6025E,
Basic
NI 6013, NI 6014
Models
Nominal Range (V)
Positive FS
Negative FS
NI 6030E, NI 6031E,
NI 6032E, NI 6033E
NI 6040E
1.43
1.201
1.405
1.176
8.127
5.685
8.127
2.417
5.685
Signal Conditioning
8.127
3.835
Table 3. E Series Analog Output Absolute Accuracy Specifications
Recommended Accessories
Signal conditioning is required for sensor measurements or voltage
inputs greater than 10 V. National Instruments SCXI is a versatile,
high-performance signal conditioning platform, intended for high-
channel-count applications. NI SCC products provide portable,
flexible signal conditioning options on a per-channel basis. Both
signal conditioning platforms are designed to increase the
performance and reliability of your DAQ System, and are up to 10X
more accurate than terminal blocks (please visit
ni.com/sigcon
for
more details). Refer to the table below for more information:
Sensor/Signals (>10 V)
System Description
DAQ Device
High performance
PCI-60xxE, PXI-60xxE, DAQPad-60xxE
Low-cost, portable
Signals (<10 V)
System Description
Shielded
Shielded
Shielded
Shielded
Shielded
Low-Cost
Low-Cost
Signal Conditioning
SCXI
SCC
Page
270
251
Ordering Information
NI PXI-6071E ................................................................777676-01
NI PCI-6071E ................................................................777515-01
NI PXI-6070E ................................................................777060-01
NI PCI-6070E ................................................................777305-01
NI DAQPad-6070E for FireWire ..........................(See page 207)
NI DAQCard-6062E ..............................................(See page 207)
NI PXI-6052E ................................................................777962-01
NI PCI-6052E ................................................................777745-01
NI DAQPad-6052E for FireWire ..........................(See page 207)
NI PXI-6040E ................................................................777484-01
NI PCI-6040E ................................................................777383-01
NI PCI-6033E ................................................................777516-01
NI PCI-6032E ................................................................777422-01
NI PXI-6031E ................................................................777636-01
NI PCI-6031E ................................................................777514-01
NI PXI-6030E ................................................................777555-01
NI PCI-6030E ................................................................777384-01
NI DAQPad-6020E for USB ..................................(See page 207)
Includes NI-DAQ driver software and calibration certificate.
DAQ Device
Terminal Block
PCI-60xxE/DAQPad-60xxE
SCB-68
PXI-60xxE
TB-2705
PCI-6071E/PCI-6033E/PCI-6031E
SCB-100
PXI-6071E/PXI-6031E
Two TBX-68s
DAQPad-60xxE
SCB-68
PCI-60xxE/PXI-60xxE/DAQPad-60xxE
CB-68LP
DAQCard-60xxE
CB-68LP
Cable
SH6868-EP
SH6868-EP
SH100100
SH1006868
SHC6868-EP
R6868
RC6868
Page
214
214
214
214
214
214
214
For more information on warranty and value-added services,
see page 20.
BUY ONLINE!
Visit
1Terminal Blocks do not provide signal conditioning (ie. filtering, amplification, isolation, etc.), which may be necessary to
increase the accuracy of your measurements.
Table 4. Recommended Accessories
201
Multifunction DAQ Overview
Calibration
DAC
CHO
Amplifier
Anti-
Aliasing
Filter
12 or 16-Bit
ADC
CHX
Amplifier
Anti-
Aliasing
Filter
12 or 16-Bit
ADC
AI FIFO
NI MITE
Bus
Interface
PCI/PXI Bus
Analog
Trigger
Circuitry
I/O Connector
PFI
Digital/Analog
Trigger
AI
Timing/Control
DMA/INT
Request
Two 24-Bit
Counter/Timers
NI
DAQ-STC
Bus
Interface
DI FIFO
DO FIFO
Digital I/O
(8)
AO Timing/
Control
RTSI/PXI
Trigger Bus
RTSI/PXI Trigger Bus
DAC 0
AO FIFO
Signal Conditioning
DAC 1
Calibration
DAC
On Selected
S Series Devices
Figure 1. S Series Hardware Block Diagram
Calibration
DAC
Analog
Input
Muxes
NI
PGIA
12 or 16-bit
ADC
AI FIFO
PCI/PXI Bus
Analog
Trigger
Circuitry
PFI
Digital/Analog
Trigger
NI MITE
Bus
Interface
AI
Timing/Control
DMA/INT
Request
Figure 2. E Series Hardware Block Diagram
I/O Connector
Two 24-bit
Counter/Timers
NI
DAQ-STC
Bus
Interface
DI FIFO
DO FIFO
Digital I/O
(8)
AO Timing/
Control
RTSI/PXI
Trigger Bus
DAC 0
AO FIFO
Calibration
DAC
On Selected E
Series Devices
DAC 1
190
12-Bit E Series Multifunction
DAQ Specifications
12-Bit E Series Specifications
Specifications – NI 607xE, NI 6062E, NI 6040E, NI 602xE
These specifications are typical for 25 °C unless otherwise noted.
Analog Input
Accuracy specifications ................................... See page 228.
Input Characteristics
6070E
6062E
6040E
602xE
6071E
Number of Channels
16 single-ended or 8 differential
(software selectable per channel)
Device
607xE
6062E
6040E
6023E
6024E
6025E
6020E
Overvoltage Protection
Powered On
Powered Off
±25 V
±15 V
±40 V
±25 V
±35 V
±25 V
64 single-ended or 32 differential
Resolution......................................................... 12 bits, 1 in 4,096
607xE
6062E
6040E
6023E
6024E
6025E
6020E
Maximum Sampling Rate
1.25 MS/s
500 kS/s
500 kS/s single-channel scanning
250 kS/s multichannel scanning
200 kS/s
6070E
6062E, 6040E
602xE
6071E
Inputs Protected
AI <0..15>, AI SENSE
AI <0..63>, AI SENSE, AI SENSE2
100 kS/s
Input Signal Ranges
Range (Software Selectable)
Bipolar Input Range
20 V
±10 V
10 V
±5 V
5V
±2.5 V
2V
±1 V
1V
±500 mV
500 mV
±250 mV
200 mV
±100 mV
100 mV
±50 mV
20 V
±10 V
10 V
±5 V
1V
±500 mV
100 mV
±50 mV
FIFO Buffer Size
DAQCard-6062E
DAQPad-6020E
DAQPad-6070E
DAQCard-6024E
PCI/PXI-6070E
6071E, 6040E
PCI-6023E, NI 6025E, PCI-6024E
8,192 samples
4,096 samples
2,048 samples
512 samples
Device
607xE
6062E
6040E
6020E
6023E
6024E
6025E
Unipolar Input Range
0 to 10 V
0 to 5 V
0 to 2 V
0 to 1 V
0 to 500 mV
0 to 200 mV
0 to 100 mV
Data transfers
PCI, PXI, DAQPad for FireWire ..................
DAQCard, DAQPad for USB.......................
DMA modes
Configuration memory size ..............................
DMA, interrupts, programmed I/O
Interrupts, programmed I/O
Scatter-gather (single-transfer, demand transfer)
512 words
Transfer Characteristics
Device
607xE
6062E
6040E
6023E
6024E
6025E
6020E
Relative Accuracy
Typical Dithered
Maximum Undithered
±0.5 LSB
±1.5 LSB
Signal Conditioning
Input coupling................................................... DC
Maximum working voltage
(signal + common mode) ........................... Input should remain within ±11 V of ground
±0.2 LSB
DNL
Typical
±0.5 LSB
±1.5 LSB
Device
607xE
6040E
6023E
PCI-6024E
6025E
6020E
6062E
DAQCard-6024E
Maximum
±1.0 LSB
±0.2 LSB
±0.75 LSB
±1.0 LSB
-0.9, +1.5 LSB
No missing codes............................................. 12 bits, guaranteed
233