首页 > 器件类别 >
PCA-121B2W
SIGNAL CONDITIONER & CONVERTERS
厂商名称:ETC
INSTALATION, OPERATION & MAINTENANCE
MANUAL
FOR
PA-1001A Series
SIGNAL CONDITIONER
& CONVERTORS
PA1001A – 7/02
Page 1 of 11
CONVERTERS
PA1001A Series
INTRODUCTION:
The PA1001A series preamplifiers/signal conditioners and converters are suitable for use
with either magnetic or RF type pickup coils and allow convenient interfacing with process
measuring devices and digital electronics.
When used with Magnetic type pickups they accept low level signals and produces a pulse
output, while providing rejection of unwanted noise and false signals.
A RF type pickup reduces the pickup drag associated with conventional magnetic pickups,
resulting in a significant increase in the usable range. The PA1001A series excite a RF type
pickup with a 20KHz signal. Motion of ferrous material is sensed and the motion modulates
the coil field and subsequent conditioning provides a pulse output signal.
The output signal of the PA1001A series have a square pulse wave form, each pulse
representative of an actuator (rotor blade, gear tooth, turbine blade, etc.) passing by the
sensing area of the pickup.
Four output signal options are available:
i.
ii.
iii.
TTL/CMOS compatible
Open collector
0 to 10VDC square wave and
A square wave that is tied to the input power
iv.
The unit is powered by a user supplied 8 to 30 VDC filtered power supply. An on-board
regulator provides the required regulation and noise rejection.
Optional features include factory installed pulse scaling and field installable plug in modules
for Analog output. Pulse scaling provides the user with scaling options of
÷2, ÷4, ÷8, ÷16,
÷32, ÷64,
or
÷128,
which is useful when a remote device has a maximum count speed that
is less than the pulse-generating device. Analog output provides a current and voltage
outputs proportional to RPM, flow rate etc.
PA1001A – 7/02
Page 2 of 11
CHARACTERISTICS AND SPECIFICATIONS:
POWER INPUT:
Voltage: 8-30 VDC
Current Draw: 125mA @ 24VDC, with all options equipped.
Reverse polarity protected
PICKUP INPUT:
RF Pickup:
Modulation Carrier Frequency: 20KHz
Input Frequency: 2 Hz to 5KHz, adjustable trigger level
Trigger Sensitivity: 20 millivolt peak to peak (minimum)
Carrier Adjustment: to allow precise tuning of carrier signal to physical environment (e.g. air gap,
wall thickness)
Maximum Air Gap: 0.160”
Preferred Coil Inductance: 1.0 milliHenry +/- 10%
Maximum Distance From Pickup To Preamplifier 100 feet
Magnetic Pickup:
Input protected, RF and band pass filtered
Input Frequency: 0 to 10KHz, adjustable trigger level
Trigger Sensitivity: 4 millivolt RMS (minimum)
NOTE: At 4mV frequency range 10 Hz to 3KHz
Input Impedance: 40K Ohms (nominal)
Maximum Input Signal: 600 Volts RMS
Maximum Distance From Pickup To Preamp: 100 feet
PULSE OUTPUTS:
TTL/CMOS (Standard); Fan-out of 5 TTL/CMOS loads
Open Collector: Maximum
OFF
state voltage 30 VDC
Maximum
ON
current 0.40 amps
0 to 10 VDC square wave
Square wave tied to input voltage
Pulse scaling (option):
÷2, ÷4, ÷8, ÷16, ÷32, ÷64,
or
÷128
user selectable.
CURRENT OUTPUT OPTION:
Accuracy +/- 0.05% of full-scale +/-200 PPM/°C.
Available range 4 to 20mA.
Output suitable for driving floating or grounded loads.
Maximum loop impedance 500 ohms.
Response time 0.3 seconds for 10 to 90 %.
VOLTAGE OUTPUT OPTION:
Available range 0 to 5VDC or 0 to 10VDC, user selectable.
Impedance less than 10 ohms.
PA1001A – 7/02
Page 3 of 11
ENVIRONMENTAL:
OPERATING:
STORAGE:
-40°C TO 85°C
-40°C TO 150°C
ENCLOSURES:
Unit is intended for mounting in a conduit box and comes with 2 ½” on center mounting holes. An
optional explosion proof or NEMA-4x enclosure is available upon request.
APPROVALS:
EN50081-1:1992 Generic Emission Standard
EN50082-2:1995 Generic Immunity Requirements
CE compliant to EMC Directive 89/336/EECfor uses in residential, commercial, light or heavy
industrial environments.
INSTALLATION AND WIRING DETAILS:
In considering the interconnections between the PA1001A and the measurement system some
attention must be given to anticipated noise sources and to the coupling of these noise sources to the
interconnecting wiring.
Noise signals may be coupled into the wiring between the pickup coil, the PA1001A, and the
electronic measuring systems. In general, utilizing a shielded, twisted pair for the interconnection
greatly reduces this coupling. The shield should be grounded on both ends of the cable.
However, even with proper interconnecting cabling, cross talk with other signal lines or power lines
may still occur and should be avoided. Physical isolation of the wiring reduces the chance of
potential problems.
The pickup coil should not be located more than 100 feet from the PA1001A preamplifier. It is
recommended that the PA1001A be installed on or near the pickup coil to assure optimum
performance.
INSTALLATION:
The Model PA1001A should be placed in a convenient location that maintains access to the unit
should repairs or readjustment is required.
Connections to the terminal block should be carefully dressed to avoid having bare wires extend pass
the screw clamp on the terminal block. This is particularly important when the PC card is mounted
within an explosion proof enclosure. Wires should be neatly dressed near bottom of enclosure to
prevent problems when cover is installed.
For compliance with EMI/RFI requirements the pre-amplifier must be installed in a metal enclosure
such as a suitable NEMA 4 (e.g. GUAC) junction box. The enclosure must be properly grounded.
Also a grounding strap is required from the enclosure to the power supply common (pin 7).
The cable entry requires 360° grounding coverage to the enclosure case. This is obtained by using
twisted pair with full braid cable and a cable fitting that provides a metal to metal cable clamping
connection. The insulation of the cables should be stripped back to allow for connection to the
terminal block and allow for the cable fitting to clamp onto the braided shield. Any open areas should
then be covered using foil tape.
In hazardous locations where metal conduit is used, 360° grounding are achieved. All cable shields
should be terminated to the enclosure-grounding stud.
PA1001A – 7/02
Page 4 of 11
Wiring For RF Pickup Coil:
1. Connect a two conductor shielded cable from the pickup to Terminals 5 and 6. Connect the
shield to Terminal 4 (see Figure 1). NOTE: It is also preferable that the pickup coil be grounded
to the shield.
2. Connect a power source, which must be within 8-30 VDC, to Terminals 7 and 8. The Plus
voltage is connected to Terminal 8 and the Return to Terminal 7 (see Figure 1).
3. Connect the Pulse output wire to Terminal 3, and connect the Pulse common to Terminal 4 (see
Figure 1).
4. Set Jumper HD2 (see Figure 3) to the RF position.
5. Configure Jumper HD1 (see Figure 3) to the desired pulse output as follows:
Pulse Output Type
Open Collector
Square Wave to Vin
TTL/CMOS
0-10 VDC Square Wave
HD1 Jumper Configuration
JP2
JP1 and JP2
JP1, JP2 and JP3
JP1, JP2 and JP4
6. Install pickup (e.g. in meter) and provide power to PA1001A. Use a digital voltmeter set to DC
voltage and attach the positive lead to Terminal 6 and the common lead to Terminal 2. Adjust the
Carrier Adjustment (R2) (see Figure 3) to a reading of 3.00 VDC to fine-tune the pickup for air
gap, wall thickness, etc.
7. With power still supplied to PA1001A determine if pulse output is being generated in the absence
of motion. Adjust the RF Trigger (R24) (see Figure 3) clock wise until the false pulsation has
stopped. If pulsation continues when adjusted fully clock wise, check that all shields have been
properly terminated and grounded.
PA1001A – 7/02
Page 5 of 11
历年电子大赛试题汇总分类分析
本帖最后由 paulhyde 于 2014-9-15 09:13 编辑 元旦三天彻底颓废了...
jishuaihu 电子竞赛
超大功率的电源并联,需要注意关注那些方面的设计?
平常我们都是一块电源模块能解决工业问题,但在一些超大设备负荷传动中,一套电源可能无法满足要求,需要多...
eeleader 工控电子
精密检测仪器市场空间广阔
精密检测仪器与我们的日常生活息息相关,但很多人对此并不甚了解。精密检测仪器成为检测工业产品必备的设...
给自己的一封信 综合技术交流
SONY ICD-SX2000录音笔拆解
ICD-SX2000是SONY的一款中低端录音笔,内部集成16G存储器,支持扩展TF...
littleshrimp 综合技术交流
利用TPS2393A实现大电流热插拔应用
利用TPS2393A实现大电流热插拔应用 ...
Jacktang 模拟与混合信号
车喇叭驱动电路,有没有更好的电路,推荐一下,这个共射电路是不是不能驱动车的喇...
车喇叭驱动电路,有没有更好的电路,推荐一下,这个共射电路是不是不能驱动车的喇叭。...
QWE4562009 电路观察室
