PS25201A / B
EPIC Ultra High Impedance
Electrophysiological Sensor
Data Sheet 291841 issue 3
Ultra high input resistance, typically 20G .
Dry-contact capacitive coupling.
Input capacitance as low as 15pF.
Lower -3dB point typically 200mHz.
Upper -3dB point typically 10kHz.
Operates with bipolar power supply from ±2.4V to
±5.5V.
Sensors supplied in a custom package with exposed
pins for surface mount assembly.
Ordering Information
PS25201B
Custom package
-25° to +75°
Plessey Semiconductors Electric Potential
Integrated Circuit (EPIC) product line targets a
range of applications.
The PS25201B is an ultra high impedance
solid state ECG (electrocardiograph) sensor. It
can be used as a dry contact ECG sensor
without the need for potentially dangerous low
impedance circuits across the heart. The
resolution available is as good as or better
than conventional wet electrodes.
The device uses active feedback techniques to
both lower the effective input capacitance of the
sensing element (Cin) and boost the input
resistance (Rin). These techniques are used to
realise a sensor with a frequency response
suitable for both diagnostic and monitoring ECG
applications.
APPLICATIONS
Contact ECG signal detection for:
Non-critical patient monitoring
equipment.
Emergency response diagnostics.
Lifestyle sports and health products.
Suitable for long-term and remote
monitoring.
Fig. 1 PS25201B - Top and Bottom
Fig. 2 Internal circuit of EPIC ECG Sensor
PLESSEY SEMICONDUCTORS LTD
TAMERTON ROAD | ROBOROUGH | PLYMOUTH | DEVON | PL6 7BQ
PS25201A/B
ELECTRICAL CHARACTERISTICS
amb
= -25° to +75°
C,Vdd/Vss ±2.4V to ±5.5V. The electrical characteristics are guaranteed by either production
test or by design and characterisation. They apply within the specified ambient temperature and supply voltage
unless otherwise stated.
Characteristics
Min.
Supply voltage
Supply current
Effective input resistance
Effective input capacitance
Voltage Gain (Av)
Coupling capacitance
Lower -3dB point
Upper -3dB point
4.0
47.5
±2.4
0.6
Value
Typ.
±5.5
0.20
52.5
3.5
Units
pF
@1kHz
pF
Hz
kHz
Conditions
Bipolar supply, Gnd=0V
Sensor to skin
Set by internal DC signal rejection
network – coupling capacitor 250pF
-3dB
0.2Hz
10kHz
Fig. 3 Typical Bode Plot for EPIC ECG Sensor
PS25201A/B
PIN ASSIGNMENT
Output
Pin 1
Vss
Pin 4
Gnd
Pin 2
Top View
Vdd
Pin 3
Fig. 4 Pin Assignment for the PS25201B
MECHANICAL DIMENSIONS
The package diagram is shown below. It is recommended that a solder pad 1.6mm diameter be defined
for the mounting of the sensor pins.
10.50 ± 0.08
3.44 ± 0.15
10.50 ± 0.08
0.75
9.0
0.75
0.51
0.10
1.35
0.35
7.80 ± 0.64
1.35
0.35
1.35 ±0.35
0.75
9.0
10.50
0.08
7.80
0.64
10.50
0.08
0.75
1.35 ±0.35
Top
Side
φ
Bottom
Fig. 5 Mechanical Drawing (all dimensions are nominal and in mm)
PS25201A/B
ELECTROSTATIC DISCHARGE (ESD) PROTECTION
The PS25201B is manufactured using a high performance analog CMOS process. As for all CMOS
components, it is essential that conventional ESD protection protocols be applied for the handling of this
device.
PATENTS
This component and many of the associated applications are covered by the following international
patents:
602 32 911.6-08 (DE)
AU2007228660
CA2646411
CN200780026584.8
EP1451595 (CH)
EP1451595 (ES)
EP1451595 (FR)
EP1451595 (IE)
EP1451595 (IT)
EP1451595 (NL)
EP2002273
EP2047284
EP2174416
GB1118970.1
JP2009-500908
JP4391823
TW097126903
TW1308066
US12/293872
US12/374359
US12/669615
US13/020890
US13/163988
US7885700
PS25201A/B
APPLICATION OF THE ECG SENSOR
Because of the large coupling capacitance to the body (around 250pF) the EPIC sensor’s internal electrometer can
be used in differential mode to recover true surface potential ECG signals from the surface of the skin. A typical
ECG signal at the surface of the skin is 1mV p-p.
Vdiff
Fig. 7 Differential measurement of body (skin) surface potential to produce ECG trace
Fig. 8 Comparison of two vectors from a pair of EPIC sensors (top)
and two conventional Ag/AgCl electrodes (bottom)