RF and Microwave Power Sensors/Meters
Tektronix PSM3000, PSM4000, and PSM5000 Series Data Sheet
Included Applications run under Microsoft Windows
Power Meter Application
High-speed Logging Application
LabVIEW Drivers and Programming Examples for Most Common
Windows Programming Environments are Available for Automated
System Support
Max Hold and Relative Measurement Modes
Offset, Frequency Response, and 75
Minimum Loss Pad Correction
Flexible Averaging Modes for Quick, Stable Measurements
TTL Trigger Input and Output allow Synchronization with External
Instruments
Features & Benefits
Key Performance Specifications
8 GHz, 18 GHz, 20 GHz, and 26.5 GHz Models
Models Available with N and 3.5 mm Connectors
Dynamic Range as Low as –60 dBm and as High as +20 dBm
Uncertainty as Low as 2.6%
Reading Rates up to 2000 Readings/s
Pass/Fail Limit Mode
Compact Size
The PSM3000 Series offers True Average Power Measurements that
Give Accurate Results Regardless of Signal Shape or Modulation
The PSM4000 and PSM5000 Series offer:
Pulse Power, Duty Cycle, Peak Power, and Crest Factor
Measurements
Measure Peak, Average and Minimum Power on Bursts with
Adjustable Offset and Duration
The PSM5000 Series includes a Pulse Profiling Application for Making
Measurements on Repetitive, Pulsed Signals
Builds and Displays a Trace of the Pulse Envelope
Full-trace and Gated Measurements including Pulse, Peak and
Average Power, Overshoot, Crest Factor, Rise and Fall Time, Pulse
Width, Pulse Repetition Frequency, Duty Cycle
Statistical Measurements on the Trace Data, such as Complementary
Cumulative Distribution Function (CCDF), and Probability Density
Function (PDF)
Meters are Calibrated over Full Temperature Range – No zero or
cal needed before making measurements, saving time and avoiding
poor-quality data
Average Power, Duty Cycle Corrected Pulse Power, and Measurement
Logging on All Models
Data Sheet
General-purpose RF and Microwave Average Power Measurements
Characterization of Repetitive Pulsed Signals, such as Navigation,
Weather, and other Radar
Peak and Average Power Measurements on Modulated Signals such as
GSM, CDMA, WCDMA, HSPA, WiMAX up to 10 MHz
Peak and Average Power Measurements of Modulated Pulsed
Communications Signals
Level Control Feedback for Signal Sources
Validation and Characterization of Power Amplifiers, Switches, and Other
RF and Microwave Components
Service, Maintenance, and Installation of DTV, Cellular, Microwave
Radio Link, and Radio Broadcast Transmitters
Verification and Calibration of Test Equipment and Systems
Figure 1 – Software presents familiar controls and measurement presentation.
test times and provide dynamic power measurement information that
was previously unavailable. An included High-speed Logging Application
provides a mechanism for getting this data into your PC for analysis.
For custom test applications, you can communicate with the sensors
using LabVIEW, or using a fully documented API. Example programs are
provided for the most commonly used development environments. The
communications library allows your program to communicate with up to
12 sensors, eliminating the need for costly switches.
Tektronix real-time signal analyzers, arbitrary waveform generators, and
oscilloscopes that use the Windows operating system can also be used to
control the PSM Series sensor/meter and give you quick access to accurate
power measurements.
To allow synchronization with other measurement equipment all models
include Trigger In and Trigger Out TTL signals.
High-speed measurements, extensive programming tools, and
synchronization features make these sensors versatile additions to your
test setup.
3-year Warranty
Capable, Compact Power Sensors/Meters
The PSM3000, PSM4000, and PSM5000 Series are compact power
sensors/meters that deliver fast, accurate RF and microwave power
measurements. A broad range of CW and pulse modulation measurements
are available, depending on the series you choose. Each meter comes
with Windows Power Meter application software for controlling the
meter, displaying readings, and recording data. The combination of the
sensor/meter and PC provides a complete solution, eliminating the need for
a separate, dedicated meter mainframe.
No Meter Mainframe Required
With the included power meter application software, familiar meter controls
are available at the click of a mouse and readings are presented right on
your PC screen. Familiar Windows pull-down menus provide additional
controls. Data is immediately available on the PC for further analysis and
documentation. The meters communicate with the PC using standard
USB 2.0 protocols and cables for plug-and-play ease of use.
Industry-leading Performance for Demanding
Designs
Tektronix power sensors/meters come fully calibrated over their entire
operating temperature range. Sensor zeroing and meter reference
calibration have been eliminated, reducing setup time and helping to
avoid inaccurate results. These meters provide accuracy you can count
on for general-purpose CW, peak, pulse, and other modulated power
measurements. Whether doing installation or maintenance on a wireless
base station, production testing, or R&D for wireless components, the PSM
Series products serve these needs with a wide dynamic range (–60 dBM to
+20 dBm) and frequencies ranging from 10 MHz up to 26.5 GHz.
Integrate High-speed Power Measurements Into Your
Testing
Tektronix PSM Series power sensors feature the industry’s fastest
measurement speed (2000 readings/s). This can significantly reduce
www.tektronix.com
RF and Microwave Power Sensors/Meters — Tektronix PSM3000, PSM4000, and PSM5000 Series
Select the Performance/Features to Meet Your
Needs
PSM3000 Series power sensors/meters provide true average power
measurements, giving accurate power measurements independent of signal
modulation and bandwidth. The PSM4000 Series delivers average power
(CW) measurements, and adds pulse and peak power measurements for
gathering basic data on pulsed RF and microwave signals. PSM5000
Series power sensors/meters provide the same measurements as the
PSM4000, and add pulse profiling capability for signal viewing and
characterization in pulsed RF and microwave systems.
Feature
PSM3000
PSM4000
PSM5000
Frequency Range
Dynamic Range
Measurement
Speed
Measurements
10 MHz to
26.5 GHz
–55 dBM to
+20 dBm
2000 readings/s
10 MHz to 20 GHz 50 MHz to 20 GHz
–60 dBm to
+20 dBM
2000 readings/s
–60 dBm to
+20 dBm
2000 readings/s
Figure 2 – Pulse profiling software enables a thorough analysis of pulse characteristics.
True Average
Power
Average (CW)
Power
Duty Cycle
Corrected Pulse
Power
Peak Power, Pulse
Power, Duty Cycle
Peak and Average
Burst Power
Measurement
Logging
Pulse Width,
Rise/Fall,
Overshoot, Droop
Time Gated
Measurements
Pulse Waveform
Display with
Markers
A Broad Range of Pulse Envelope
Measurements
Tektronix PSM5000 Series products feature an easy-to-use, high
performance, pulse profiling, pulse (modulation), and CW power meter and
sensor in one. The PSM5000 Series product is specifically designed for
applications requiring time domain analysis of constant-envelope, repetitive
pulsed signals. It performs time domain pulse measurements such as
rise/fall time, overshoot, and droop that have typically required costly signal
analyzers. PSM5000 Series sensors use an equivalent time-sampling
technique to reconstruct repetitive, pulsed input signals. Repetitive pulses
with video bandwidths up to 10 MHz can be measured with an effective
sample rate of up to 48 MS/s.
Performance You Can Count On
In addition to industry-leading service and support, every PSM Series power
sensor/meter comes backed with a three-year standard warranty.
www.tektronix.com
Data Sheet
Electrical Specifications
All specifications apply over the full instrument operating temperature range unless
otherwise noted, after a 20 minute warm-up interval.
PSM3000 Series USB Power Meters (True Average)
Characteristic
PSM3110
PSM3120
PSM3310
PSM3320
PSM3510
Input Connector
Frequency Range
Dynamic Range
Video Bandwidth
Total Accuracy*
Calibration Factor
Uncertainty (CF)
Linearity Uncertainty (L)
Noise Uncertainty (N)
3.5 mm, male
N-type, male
10 MHz to 8 GHz
3.5 mm, male
N-type, male
10 MHz to 18 GHz
–55 dBm to +20 dBm
100 Hz, typical
3.5 mm, male
10 MHz to 26.5 GHz
10 MHz to 1 GHz: 2.5%
1 GHz to 8 GHz: 2.4%
Total Uncertainty = 2 ×
(CF/2)
+ (L/2)
+ (N/2)
+ (Z/√2)
+ (Mm/√2)
+ (T/√2)
10 MHz to 1 GHz: 1.8%
1 GHz to 8 GHz: 1.7%
1 GHz to 10 GHz: 2.4%
1 GHz to 10 GHz: 1.7%
10 GHz to 18 GHz: 2.7%
10 GHz to 18 GHz: 1.9%
18 GHz to 26.5 GHz: 3.7%
Zero Offset Power*
(Z)
Match*
+15 dBm to +20 dBm: 3.0%
–15 dBm to +15 dBm: 2.5%
–55 dBm to –15 dBm: 2.0%
5 second integration
+10 dBm to +20 dBm: 0.10%
–15 dBm to +10 dBm: 0.25%
–30 dBm to –15 dBm: 0.10%
–40 dBm to –30 dBm: 0.25%
–50 dBm to –40 dBm: 1.50%
–55 dBm to –50 dBm: 4.50%
[(3.0 nW at 25 °C) + |ΔT| × (0.15 nW / °C)] + 0.01 nW / month
1.20:1 VSWR (21 dB Return Loss)
10 MHz to 10 GHz: 1.20:1 VSWR (21 dB Return Loss)
10 GHz to 18 GHz: 1.29:1 VSWR (18 dB Return Loss)
10 MHz to 10 GHz:
1.20:1 VSWR (21 dB
Return Loss)
10 GHz to 26.5 GHz:
1.29:1 VSWR (18 dB
Return Loss)
Temperature Uncertainty
(T)
40 °C < T
50 °C: 2.00%
30 °C < T
40 °C: 0.75%
20 °C < T
30 °C: 0.00%
10 °C < T
20 °C: 0.75%
0 °C < T
10 °C: 2.00%
Total uncertainty includes contributions from Calibration Factor Uncertainty (CF), Linearity Uncertainty (L), Noise Uncertainty (N), Zero Offset Uncertainty (Z), Mismatch Uncertainty, and Temperature Uncertainty (T). All
error terms must be converted to percentages for Total Uncertainty (RSS) calculation. Mismatch Uncertainty (Mm) requires knowledge of source match and should be expressed as a percentage using the equation
Mm = 100 × [(1 ±
Γ
source
Γ
sensor
– 1].
Use the following formula to determine Zero Offset Uncertainty as a percentage: Z = (Zero Offset Power / Nominal Power) × 100.
Nominal impedance = 50
Ω.
www.tektronix.com
PSM4000 Series USB Power Meters (Average/Peak/Pulse)
Characteristic
PSM4110
PSM4120
PSM4320
PSM4410
Input Connector
Frequency Range
Dynamic Range
Maximum Peak-to-Average Ratio
Internal Video Bandwidth
Timebase
Sample Rate
Average Power, Minimum Pulse
Width
Peak Power, Minimum Pulse
Width
Total Accuracy*
(CF)
3.5 mm, male
N-type, male
10 MHz to 8 GHz
10 MHz to 8 GHz
10 MHz to 6 GHz: –60 dBm to +20 dBm
6 GHz to 8 GHz: –50 dBm to +20 dBm
10 MHz to 6 GHz: 80 dB
6 GHz to 8 GHz: 70 dB
N-type, male
3.5 mm, male
50 MHz to 18.6 GHz
50 MHz to 20 GHz
–40 dBm to +20 dBm
55 dB
10 MHz, typical
±50 ppm, typical
500 kS/s
500 ns, typical
200 ns, typical
(CF/2)
+ (L/2)
+ (N/2)
+ (Z/√2)
+ (Mm/√2)
+ (T/√2)
50 MHz to 500 MHz: 4.0%
10 MHz to 100 MHz: 7.0%
500 MHz to 12.5 GHz: 2.6%
500 MHz to 10 GHz: 1.7%
100 MHz to 500 MHz: 4.0%
12.5 GHz to 18 GHz: 3.2%
10 GHz to 18.6 GHz: 1.9%
500 MHz to 8 GHz: 1.7%
500 MHz to 8 GHz: 2.5%
18 GHz to 20 GHz: 3.5%
50 MHz to 100 MHz
10 MHz to 100 MHz
+15 dBm to +20 dBm: 7.0%
–40 dBm to +15 dBm: 5.0%
+10 dBm to +15 dBm: 5.0%
–60 dBm to +10 dBm: 4.0%
100 MHz to 2 GHz
100 MHz to 2 GHz
+5 dBm to +15 dBm: 5.0%
–40 dBm to +5 dBm: 3.0%
–60 dBm to +10 dBm: 3.0%
2 GHz to 20 GHz
+15 dBm to +20 dBm: 6.0%
2 GHz to 8 GHz
+5 dBm to +15 dBm: 4.0%
+15 dBm to +20 dBm: 5.0%
–40 dBm to +5 dBm: 2.0%
+10 dBm to +15 dBm: 3.0%
–60 dBm to +10 dBm: 2.0%
1 second integration
+10 dBm to +20 dBm: 1.5% (50 MHz to 20 GHz)
+10 dBm to +20 dBm:
–20 dBm to +10 dBm: 1.0% (50 MHz to 20 GHz)
0.22% (10 MHz to 100 MHz)
–30 dBm to –20 dBm: 1.5% (50 MHz to 20 GHz)
0.15% (100 MHz to 8 GHz)
–40 dBm to –30 dBm: 7.0% (50 MHz to 18.6 GHz)
–30 dBm to +10 dBm:
0.04% (100 MHz to 8 GHz)
–50 dBm to –30 dBm:
0.04% (100 MHz to 6 GHz)
0.15% (6 GHz to 8 GHz)
–60 dBm to –50 dBm:
0.44% (10 MHz to 100 MHz)
0.15% (100 MHz to 6 GHz)
[(0.35 nW at 25 °C) + |ΔT| × (0.025 nW / °C)] + 0.005 nW / month
Zero Offset Power*
(Z)
50 MHz to 500 MHz
[(200 nW at 25 °C) + |ΔT| × (10 nW / °C)] + 10 nW / month
500 MHz to 20 GHz
[(100 nW at 25 °C) + |ΔT| × (5 nW / °C)] + 5 nW / month
50 MHz to 10 GHz: 1.20:1 VSWR 50 MHz to 10 GHz: 1.20:1 VSWR
(21 dB Return Loss)
(21 dB Return Loss)
10 GHz to 18.6 GHz:
1.29:1 VSWR
(18 dB Return Loss)
10 GHz to 20 GHz: 1.29:1 VSWR
(18 dB Return Loss)
Match*
1.09:1 VSWR
(27 dB Return Loss)
1.15:1 VSWR
(23 dB Return Loss)
www.tektronix.com