MS4525DO
PCB Mounted Digital Output Transducer
Combination Temperature and Pressure
C or SPI Protocol
Differential, Gage, Absolute &
Compound, Vacuum
Temperature Compensated
3.3 or 5.0 Vdc Supply Voltage
Low Power Option Available
(standby < 1uA)
The MS4525DO is a small, ceramic based, PCB mounted pressure transducer from Measurement
Specialties. The transducer is built using Measurement Specialties’ proprietary UltraStable™ process and the
latest CMOS sensor conditioning circuitry to create a low cost, high performance Digital Output Pressure (14bit)
and Temperature (11bit) transducer designed to meet the strictest requirements from OEM customers.
The MS4525DO is fully calibrated and temperature compensated with a total error band (TEB) of less
than 1.0% over the compensated pressure range. The sensor operates from single supply of either 3.3 or
5.0Vdc and requires a single external component for proper operation.
The rugged ceramic transducer is available in side port, top port, and manifold mount and can measure
absolute, gauge, differential, vacuum or compound pressure from 1 to 150psi. The 1/8” barbed pressure ports
mate securely with 3/32” ID tubing.
PSI Pressure Ranges
PCB Mountable
Digital Output
Barbed Pressure Ports
APPLICATIONS
Factory Automation,
Vacuum Switch
Altitude and Airspeed Measurements
Medical Instruments
Leak Detection
STANDARD RANGE (PSI)
Pressure
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP
Absolute
Gauge
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
Differential
DS, SS,TP, MM
DS, SS,TP, MM
DS, SS,TP ,MM
DS, MM
DS, MM
DS, MM
DS, MM
DS, MM
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP, DS
Compound
Vacuum
See Package Configurations: DS= Dual Side Port, SS= Single Side Port, TP= Top Port, MM= Manifold Mount
Only
Protocol is Available on “L” type Pin Styles; Reference Ordering Information for Details
Pin Style “L” is only available SS and MM port types.
MS4525DO
www.meas-spec.com
1/14
October 2013
MS4525DO
BLOCK DIAGRAM
ABSOLUTE MAXIMUM RATINGS
Supply Voltage
Output Current
Storage Temperature
Humidity
Overpressure
Burst Pressure
ESD
Solder Temperature
Conditions
= 25 °C
= 25°C
= 25°C
= 25 °C, both Ports
= 25 °C, Port 1
HBM
Min
2.7
+125
95
Not to Exceed 300
-4
+4
250°C, 5 sec max.
%RH
psi
psi
kV
Non Condensing
See Table 1
EN 61000-4-2
TABLE 1- BURST PRESSURE BY RANGE AND PACKAGE STYLE
Range
001
002
005
015
030
050
DS
90
SS, TP, MM
psi
psi
psi
psi
psi
psi
psi
psi
MS4525DO
www.meas-spec.com
2/14
October 2013
MS4525DO
ENVIRONMENTAL SPECIFICATIONS
Mechanical Shock
Mechanical Vibration
Thermal Shock
Life
MTTF
Conditions
Mil Spec 202F, Method 213B, Condition C, 3 Drops
Mil Spec 202F, Method 214A, Condition 1E, 1Hr Each Axis
100 Cycles over Storage Temperature, 30 minute dwell
1 Million FS Cycles
>10Yrs, 70 ºC, 10 Million Pressure Cycles, 120%FS
Pressure
PERFORMANCE SPECIFICATIONS
Supply Voltage : 5.0V or 3.3 Vdc
Reference Temperature: 25°C (unless otherwise specified)
PARAMETERS
Pressure Accuracy
Total Error Band (TEB)
Temperature Accuracy
Supply Current
Compensated Temperature
Output Pressure Resolution
Output Temperature Resolution
Update Time
Start Time to Data Ready
Weight
Media
-10
-25
MIN
-0.25
-1.0
+105
TYP
MAX
0.25
UNITS
%Span
%Span
ºC
ºC
ºC
bits
bits
mS
mS
grams
NOTES
Non-Corrosive Dry Gases Compatible with Ceramic, Silicon, Pyrex,
PPS, RTV, Gold, Aluminum and Epoxy. See “Wetted Material by Port
Designation” chart below.
1. Proper operation requires an external capacitor placed as shown in Connection Diagram. Output is not ratiometric to supply voltage.
2. The maximum deviation from a best fit straight line (BFSL) fitted to the output measured over the pressure range at 25C. Includes all
errors due to pressure non linearity, hysteresis, and non repeatability.
3. Total pressure error band includes all accuracy errors, thermal errors over the compensated temperature range and span and offset
calibration tolerances. For ideal sensor output with respect to input pressure and temperature, reference Transfer Function charts
below. TEB values are valid only at the calibrated supply voltage.
4. The deviation from a best fit straight line (BFSL) fitted to the output measured over the compensated temperature range.
5. For errors beyond the compensated temperature range, see Extended Temperature Multiplier chart below.
6. Start time to data ready is the time to get valid data after POR (power on reset). The time to get subsequent valid data is then specified
by the update time specification.
7. This product can be configured for custom OEM requirements, contact factory for lower power consumption or higher accuracy.
MS4525DO
www.meas-spec.com
3/14
October 2013
MS4525DO
CONNECTION DIAGRAM
1.
Place 100nF capacitor between Supply and GND to within 2 cm of sensor.
PRESSURE AND TEMPERATURE TRANSFER FUNCTION
Gage, Differential and Compound Pressure Types
3FFF
3CCB
399A
7FF
733
666
Output( 2 HEX counts)
599
4CC
266
199
CC
2000
0333
0000
pe
Ty
pe
Ty
Pmax
Pmin
Pressure
-50 °C
Tmin
-10 °C
85 °C
150 °C
Tmax
A Type:
B Type:
Output( Decimal counts) =
80% *16383
__________
* (Pressure
applied
-Pmin) + 10%*16383
Pmax-Pmin
90%*16383
__________
Pmax-Pmin
* (Pressure
applied
-Pmin) + 5%*16383
Temperature
Output (Decimal Counts)
(Output °C - (-50°C)
Tmin
) * 2047
_________________________________
(150°C
Tmax
- (-50°C)
Tmin
MS4525DO
www.meas-spec.com
4/14
October 2013
MS4525DO
Vacuum Pressure Type
3FFF
3CCB
399A
7FF
733
AT
ype
BT
ype
666
599
4CC
266
199
2000
0333
0000
CC
Pmin(-15 PSI G)
Pressure
Pmax(0 PSI G)
-50 °C
Tmin
-10 °C
85 °C
150 °C
Tmax
A Type:
B Type:
80% *16383
__________
* (Pmax- Pressure
applied
) + 10%*16383
90%*16383
__________
* (Pmax- Pressure
applied
) + 5%*16383
Temperature
Tmin
) * 2047
(150°C
Tmax
- (-50°C)
Tmin
MS4525DO
www.meas-spec.com
5/14
October 2013