The MBC201 Series of open-frame medical power supplies, with its
wide universal input range 90-264 VAC, is available at 200 W of
output power with forced-air cooling and a variety of single output
voltages.
The MBC Series is designed and approved to the latest medical
standards (EN/IEC 60601-1), providing 2 x MOPP isolation for Class I
& Class II applications.
These power supplies are ideal for medical, telecom, datacom,
industrial equipment and other applications.
5 x 3 x 1.5 Inch Form Factor
200 W with Forced-Air Cooling
Approved to EN/IEC 60601-1
High Efficiency > 88%
12 V Fan & 5 V Standby Outputs
Remote Sense
IEC Protection Class Options:
Class I: Earthing Tab J4 (no suffix)
Class II: No Earthing Tab (-2 suffix)
Medical Safety Agency Approvals
RoHS Compliant
CE Marked
Diagnostic
Drug Pump
Monitoring
Dialysis
Home Health Care
Portable Equipment
MBC201 Series
MODEL
MBC201-1005G
MBC201-1T05G
MBC201-1012G
MBC201-1T12G
MBC201-1015G
MBC201-1T15G
MBC201-1024G
MBC201-1T24G
MBC201-1030G
MBC201-1T30G
MBC201-1048G
MBC201-1T48G
COVER-201-XBC
CONNECTOR
JST
Screw Terminal
JST
Screw Terminal
JST
Screw Terminal
JST
Screw Terminal
JST
Screw Terminal
JST
Screw Terminal
OUTPUT VOLTAGE
5.0 VDC
5.0 VDC
12 VDC
24
48
MAX LOAD
CONVECTION
26.0 A
26.0 A
13.33 A
13.33 A
10.66 A
10.66 A
6.66 A
6.66 A
5.33 A
5.33 A
3.33 A
3.33 A
MAX L2AD
300 LFM
2 3, 4
26.0 A
35.0 A
16.67 A
16.67 A
13.33 A
13.33 A
8.33 A
8.33 A
6.66 A
6.66 A
4.17 A
4.17 A
MINIMUM LOAD
0.0 A
0.0 A
0.0 A
0.0 A
0.0 A
0.0 A
0.0 A
0.0 A
0.0 A
0.0 A
0.0 A
0.0 A
RIPPLE &
NOISE
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
Metal cover kit accessory
Specifications are for nominal input voltage, 25°C unless otherwise stated.
PARAMETER
Input Voltage
Input Frequency
Input Current
Inrush Current
Leakage Current
Power Factor
Switching Frequency
120 VAC:
230 VAC:
120 VAC:
230 VAC:
120 VAC:
230 VAC:
120 VAC:
230 VAC:
PFC converter (variable)
Resonant converter (variable)
DESCRIPTION / CONDITION
Universal
SPECIFICATION
90-264 VAC / 120-390 VDC
47 to 63 Hz
2.4 A max.
1.2 A max.
35 A max.
65 A max.
< 150 µA
< 300 µA
0.99
0.95
35 to 250 kHz, 90 kHz typical
For Class II (without input Earth pin) add suffix -2 (e.g.: MBC201-1012G-2).
Combined output power from V1, VSTBY and VFAN should not exceed the total output power rating.
Fan output voltage tolerance is +/-20%. During V1 full load, VFAN needs min. 20 mA load to be within regulation band.
Peak current for fan output is 1 A.
Ripple is 2% up to 20% load and < 1% above 20% load. Ripple is peak to peak with 20 MHz bandwidth and 10 µF (Tantalum
capacitor) in parallel with a 0.1 µF capacitor at rated line voltage and load ranges.
When used in Cover Kit, de-rate output power to 70 % under all operating conditions.
tech.support@psbel.com
MBC201 Series
PARAMETER
Efficiency
Hold Up Time
Output Power
Line Regulation
Load Regulation
Transient Response
Rise Time
Set Point Accuracy
Voltage Adjustment
Over Voltage Protection
Over Current Protection
Short Circuit Protection
Short term, auto recovery
V1
V1 only
Main output 50 to 100% load change,
50 Hz, 50% duty cycle, 0.1A / µs
120 VAC:
230 VAC:
120 / 230 VAC
SPECIFICATION
84% typical
86% typical
10 ms
160 W to 200 W
+/-0.5%
+/-2.0%
< 10%,
recovery time < 5 ms
< 100 ms
± 1%
±3%
110 to 150 %
110% typical above rating
Derate output power linearly to 80% from 90 V to 80 VAC input.
Peak Power: 250 W for 0.2 s
PARAMETER
Storage Temperature
Refer to derating curve (
Figure 1
Start-up is guaranteed
Convection:
300 LFM:
Convection:
300 LFM:
SPECIFICATION
-20 to 70°C
-20 to 0°C
-40 to 70° C
130 W
175 W
160 W
200 W
95%
10,000 ft.
40,000 ft.
1.6 million hours
5 V model
Cooling
Other models
Humidity
Altitude
Reliability
Non Condensing
Operating:
Non-Operating:
MTBF according to Telcordia
–SR332-Issue
Conducted Emissions
Radiated Emissions
Input Current Harmonics
Voltage Fluctuation and Flicker
ESD Immunity
Radiated Field Immunity
Electrical Fast Transient Immunity
Surge Immunity
Conducted Immunity
Magnetic Field Immunity
Voltage Dips, Interruptions
EN 55011-B, CISPR22-B, FCC PART15-B
EN 55011 A;
with external core (King core K5B RC 25x12x15-M in input cable)
EN 61000-3-2
EN 61000-3-3
EN 61000-4-2
EN 61000-4-3
EN 61000-4-4
EN 61000-4-5
EN 61000-4-6
EN 61000-4-8
EN 61000-4-11
Pass
Pass
Level B
Class D
Pass
Level 4, Criterion A
Level 3, Criterion A
Criterion A & B
Asia-Pacific
+86 755 298 85888
Europe, Middle East
+353 61 225 977
North America
+1 408 785 5200
BCD.00289_AF
© 2017 Bel Power Solutions & Protection
MBC201 Series
Input to Output:
Approved to the latest edition of the following standards:
CSA/UL60601-1, EN60601-1 and IEC60601-1.
Nemko, UL, C-UL
Complies with LVD Directive
PARAMETER
Isolation Voltage
Safety Standards
Agency Approvals
CE mark
SPECIFICATION
Min. 5900 VDC
PARAMETER
Power Good Signal
Power Fail Signal
Remote ON / OFF
Remote Sense
TTL signal goes high after main output is in regulation band. Delay is 0.1 to 0.3 sec.
TTL signal goes low 1 ms advance before output goes out of regulation due to mains failure
To turn-on power supply short J3 pin 4 to pin 6 (PSU is shipped in this configuration)
Compensates for 200 mV cable drop
CONNECTOR
AC Input Connector
PIN
J1
Pin 1
AC Neutral
Pin 2
AC Line
Pin 1,2,3
Pin 4,5,6
Pin 1
Pin 2
Pin 3
Pin 4
Pin 5
Pin 6
Pin 7
Pin 8
RTN
Vout
+VE Remote Sense
VFAN (+12 V/0.5 A)
-VE Remote Sense
Remote ON/OFF
VSTBY (+5 V/1 A, +/-5%)
RTN (Signal)
Power Fail
Power Good
DC Output Connector
J2
MANUFACTURER / PN
Molex: 26-60-4030 or equivalent;
Mating: 09-50-3031; Pins: 08-50-0106
Option 1: Tyco: 2-1776112-3 or equivalent
Mating: 13 AWG wire
Option 2: JST: B6P-VH-B (LF) (SN) or B6P-VH (LF) (SN) or
equivalent
Mating: VHR-6M; Pins: SVH-41T-P1.1
Signals & Aux Power
J3
Molex: 22-23-2081 or equivalent
Mating: 22-01-2087, Pins: 08-50-0113
Earthing Tab
J4
Molex: 19705-4301 or equivalent
Mating: 190030001
PARAMETER
Weight
Dimensions
Cooling
325 g (0.72 lbs.)
127.0 x 76.2 x 38.1 mm (5 x 3 x 1.5 inch)
Convection: 83 W; 300 LFM: 175 W (5 V model)
Convection: 160 W; 300 LFM: 200 W (other models)
PSU is supplied with J3 housing, pin-4 and pin-6 shorted to enable main output without remote on/off feature
MBC201 Series
Convection load: 83 W up to 40 °C
De-rate above 40 °C @ 1.89% per °C
Forced air cooled load: 130 W up to 50°C
(for screw terminal version)
De-rate above 50 °C @ 1.63% per °C
Forced air cooled load: 175 W up to 50°C
(for JST connector version)
De-rate above 50 °C @ 2.5% per °C
Convection load: 160 W up to 40 °C
De-rate above 40 °C @ 2% per °C
Forced air cooled load: 200 W up to 50°C
Figure 1. Derating Curves
Asia-Pacific
+86 755 298 85888
Europe, Middle East
+353 61 225 977
North America
+1 408 785 5200
BCD.00289_AF