首页 > 器件类别 > 电源/电源管理 > 电源电路
REM3.5E-2412D/R8/A/X1 (51001639)
Regulated DC/DC Converter; In-V: 18-36; Out-V: 12; In V MIN: 18; In V MAX: 36; Out V MIN: 12; Out V MAX: 12; Out-mA: +/-145; Isolation: 8kV; Output: dual; Power Watts or Amps: 3.5W; Package: DIP24; Short Circuit Protec: Yes; MC INPUT-V: 24; MC OUTPUT-V: 12; Length: 31.8 mm; Width: 20.3 mm; Height: 10.2 mm; Packaging Type: Tube
器件类别:电源/电源管理    电源电路
厂商名称:RECOM
器件标准:
是否无铅
不含铅
是否Rohs认证
符合
Objectid
7247240106
包装说明
compliant
ECCN代码
EAR99
模拟集成电路 - 其他类型
DC-DC REGULATED POWER SUPPLY MODULE
Regulated
Converter
2MOPP, 250VAC working voltage isolation
Clearance and creepage distance >8mm
Up to 10kVDC reinforced insulation
IEC/EN/UL 60601 certified with CB Report
(3rd Ed. Safety, 4th Ed. EMC)
-40°C to +85°C operation, no derating
2:1 wide input range
DC/DC Converter
REM3.5E
3.5 Watt
2:1 Input
DIP24
Single & Dual
Output
The REM3.5E series of medical grade regulated DC/DC converters feature reinforced 250VAC
continuous working isolation with >8mm creepage/clearance. The compact DIP24 package offers
industry standard pinouts with tightly regulated single/dual outputs and UVLO, SCP, OCP and OVP.
The operating ambient temperature range is from -40°C to +85°C without derating. The converters are
UL marked and certified to CB, IEC, EN and ANSI/AAMI 60601 3rd. Ed. Safety and 4th Ed. EMC medical
standards. The low 1µA leakage current makes them suitable for medical B, BF and CF applications.
Selection Guide
Part
Number
REM3.5E-xx05S/R
(3)
/A
(4,5)
REM3.5E-xx09S/R
(3)
/A
(4,5)
REM3.5E-xx12S/R /A
(3)
(3)
(4,5)
(4,5)
nom. Input
[VDC]
5 / 12 / 24 / 48
5 / 12 / 24 / 48
5 / 12 / 24 / 48
5 / 12 / 24 / 48
5 / 12 / 24 / 48
5 / 12 / 24 / 48
5 / 12 / 24 / 48
5 / 12 / 24 / 48
5 / 12 / 24 / 48
Output
[VDC]
±5
±9
±12
±15
Output
Current
[mA]
700
388
290
233
145
±350
±194
±145
±117
Efficiency
typ.
76 / 80 / 81 / 82
80 / 81 / 82 / 82
82 / 82 / 83 / 82
83 / 82 / 84 / 83
82 / 82 / 84 / 83
76 / 80 / 81 / 82
80 / 81 / 82 / 82
82 / 82 / 83 / 82
83 / 82 / 84 / 83
Max. Capacitive
Load
(2)
[µF]
4700
3300
2200
2200
±2200
±1600
±1000
±1000
2MOPP
250VAC
REM3.5E-xx15S/R /A
REM3.5E-xx24S/R
(3)
/A
(4,5)
REM3.5E-xx05D/R
(3)
/A
(4,5)
REM3.5E-xx09D/R
(3)
/A
(4,5)
REM3.5E-xx12D/R
(3)
/A
(4,5)
REM3.5E-xx15D/R
(3)
/A
(4,5)
Notes:
Note1: Efficiency is tested at nominal input and full load at +25°C ambient
Note2: Max Cap Load is tested at nominal input and full resisitive load
E314885
CAN/CSA-C22.2 No. 60601-1:14
ANSI/AAMI ES60601-1
EN60601-1 pending
IEC60601-1 pending
IEC60601-1-2 pending
EN55032 pending
Model Numbering
REM3.5E-
Output Power
nom. Input Voltage
Output Voltage
Single
or
Dual
Notes:
Note3:
Note4:
Note5:
Note6:
Ordering Examples
REM3.5E-0505S/R8/A
REM3.5E-1205D/R10/A/ CTRL
REM3.5E-2405S/R8/A/X1
REM3.5E-2405D/R10/A/CTRL/X1
/R /A /
UVLO Option
(5)
CTRL Option
(4)
Pinning
Isolation
(3)
add suffix „/R8“ for 8kVDC or „/R10“ for 10kVDC isolation
add suffix „/CTRL“ for fitted CTRL pin
add suffix „/X1“ for Under Voltage Lockout Option
SMD versions available from Q2/2019
= 5Vin, 5Vout, Single, 8kVDC Isolation and „A“ pinning, DIP24
= 12Vin, 5Vout, Dual, 10kVDC Isolation, „A“ pinning, with CTRL pin
= 24Vin, 5Vout, Single, 8kVDC Isolation, „A“ pinning, DIP24 and with UVLO Option
= 24Vin, 5Vout, Dual, 10kVDC Isolation, „A“ pinning, DIP24, CTRL pin and UVLO Option
www.recom-power.com
REV.: 0/2019
MED-1
REM3.5E
DC/DC Converter
(measured @ Ta= 25°C, nom. Vin, full load and after warm-up unless otherwise stated)
BASIC CHARACTERISTICS
Internal Input Filter
Input Voltage Range
nom. Vin = 5VDC
nom. Vin = 12VDC
nom. Vin = 24VDC
nom. Vin = 48VDC
nom. Vin= 5VDC
Under Voltage Lockout (UVLO)
("/X1" version)
nom. Vin= 12VDC
nom. Vin= 24VDC
nom. Vin= 48VDC
DC-DC ON
DC-DC OFF
DC-DC ON
DC-DC OFF
DC-DC ON
DC-DC OFF
DC-DC ON
DC-DC OFF
4.5VDC
9VDC
18VDC
36VDC
5VDC
12VDC
24VDC
48VDC
3.9VDC
9VDC
7.9VDC
18VDC
16.7VDC
36VDC
34.3VDC
900mA
360mA
180mA
90mA
50mA
20mA
5mA
2.5mA
10%
0.6ms
0.45ms
0.6ms
Open or 0VDC<V
CTRL
<1.2VDC
Short or 4.8VDC<V
CTRL
<12VDC
25mA
350µA
120kHz
20MHz BW
Notes:
Note7: Measurements are made with a 0.1µF MLCC across output. (low ESR)
150mVp-p
Condition
Min.
Typ.
Pi-type
9VDC
18VDC
36VDC
75VDC
4.5VDC
Input Current
nom. Vin = 5VDC
nom. Vin = 12VDC
nom. Vin = 24VDC
nom. Vin = 48VDC
nom. Vin = 5VDC
nom. Vin = 12VDC
nom. Vin = 24VDC
nom. Vin = 48VDC
Quiescent Current
Minimum Load
(8)
Start-up time
Rise time
Hold-up time
ON/OFF CTRL
Input Current of CTRL Pin
Standby Current
Intvernal Operating Frequency
Output Ripple and Noise
(7)
DC-DC ON
DC-DC OFF
CTRL
=5VDC
DC-DC OFF
ON/OFF CTRL Option
+V
in
R1
CTRL
+V
out
R1
470Ω
-V
out
-V
in
continued on next page
www.recom-power.com
REV.: 0/2019
MED-2
REM3.5E
DC/DC Converter
Efficiency vs. Output Load
REM3.5E-0505S
REM3.5E-2405S
Efficiency [%]
4.5VDC
5VDC
7VDC
9VDC
Efficiency [%]
18VDC
20VDC
24VDC
30VDC
36VDC
Output Load [%]
Output Load [%]
Efficiency vs. Input Voltage
(@ full Load)
REM3.5E-0505S
REM3.5E-2405S
Efficiency [%]
5.0
5.5
6.0
6.5
7.0
7.5
8.0
8.5
Efficiency [%]
24
26
28
34
36
Input Voltage [VDC]
Input Voltage [VDC]
REGULATIONS
Output Accuracy
Line Regulation
Load Regulation
(8)
Cross Regulation
Transient Response
low line to high line, full load
10% to 100% load
dual output only
25% load step change
Notes:
Note8: Operation below 10% load will not harm the converter, but specifications may not be met
www.recom-power.com
REV.: 0/2019
MED-3
Condition
Value
±1.5% typ.
±0.3% max.
0.5% typ.
±5.0% max.
5ms
REM3.5E
DC/DC Converter
Deviation vs. Load
REM3.5E-0505S
REM3.5E-2405S
Deviation [%]
-1.0
-1.5
Deviation [%]
-1.0
-1.5
Output Load [%]
Output Load [%]
PROTECTIONS
Short Circuit Protection (SCP)
“/R8” suffix
below 100mΩ
I/P to O/P
I/P to O/P
tested for 1 second
rated for 1 minute
tested for 1 second
rated for 1 minute
Value
continuous, hiccup mode, automatic recovery
8kVDC
4kVAC/60Hz
10kVDC
5kVAC/60Hz
10GΩ min.
Isolation Voltage
(9)
“/R10” suffix
Isolation Resistance
Isolation Capacitance
Insulation Grade
Leakage Current
Means of Protection
Medical Device Classification
Internal
External
clearance/creepage
clearance/creepage
Notes:
Note9: For repeat Hi-Pot testing, reduce the time and/or the test voltage
Note10: Refer to local safety regulations if input over-current protection is also required. Recommended fuse: slow blow type
20pF typ.
reinforced
0.8µA typ. / 1µA max.
2MOPP
built-in power supply
>8mm
>8mm
ENVIRONMENTAL
Maximum Case Temperature
Temperature Coefficient
Thermal Impedance
Operating Altitude
Operating Humidity
Pollution Degree
MTBF
according to MIL-HDBK-217F, G.B.
www.recom-power.com
REV.: 0/2019
MED-4
Condition
full load @ natural convection 0.1m/s (see graph)
Value
-40°C to +85°C
+105°C
±0.02%/K typ. / ±0.05%/K max.
0.1m/s, horizontal
non-condensing
+25°C
+85°C
20K/W
3000m
5% - 95% RH max.
PD2
3600 x 10
hours
450 x 10
hours
REM3.5E
DC/DC Converter
Derating Graph
(@ Chamber and natural convection 0.1m/s)
Output Load [%]
-40 -30 -20 -10 0 10 20 30 40 50 60 70 80 90 100 110
85 95
Ambient Temperature [°C]
SAFETY AND CERTIFICATIONS
Certificate Type (Safety)
Medical Electric Equipment, General Requirements for Safety and
Essential Performance
Essential Performance (CB Scheme)
RoHS 2+
Report / File Number
E314885
pending
pending
Standard
CAN/CSA-C22.2 No. 60601-1:14, 3rd Edition: 2014
ANSI/AAMI ES60601-1:2012
EN60601-1:2006 + A12:2014
IEC60601-1:2005, 3rd Edition + AM1:2012
RoHS 2011/65/EU + AM2015/863
EMC Compliance
Medical electrical equipment Part 1-2: Electromagnetic
disturbances – Requirements and tests
Information technology equipment - Radio disturbance
characteristics - Limits and methods of measurement
ESD Electrostatic discharge immunity test
Radiated, radio-frequency, electromagnetic field immunity test
Fast Transient and Burst Immunity
Surge Immunity
Immunity to conducted disturbances, induced by radio-frequency fields
Condition
pending
with external filter
Air ±15kV, Contact ±8kV
10V/m
DC Power Port: ±2kV
DC Power Port: ±1kV
10Vr.m.s
Standard / Criterion
IEC60601-1-2
EN55032, Class A and B
EN61000-4-2, Criteria A
EN61000-4-3, Criteria A
EN61000-4-4, Criteria A
EN61000-4-5, Criteria A
EN61000-4-6, Criteria A
www.recom-power.com
REV.: 0/2019
MED-5
参数对比
与REM3.5E-2412D/R8/A/X1 (51001639)相近的元器件有:REM3.5E-0512D/R8/A/X1 (51001615)、REM3.5E-1212D/R8/A/X1 (51001627)、REM3.5E-4812D/R8/A/X1 (51001651)、REM3.5E-1215D/R8/A/X1、REM3.5E-1215S/R8/A/X1 (51001435)。描述及对比如下:
型号 REM3.5E-2412D/R8/A/X1 (51001639) REM3.5E-0512D/R8/A/X1 (51001615) REM3.5E-1212D/R8/A/X1 (51001627) REM3.5E-4812D/R8/A/X1 (51001651) REM3.5E-1215D/R8/A/X1 REM3.5E-1215S/R8/A/X1 (51001435)
描述 Regulated DC/DC Converter; In-V: 18-36; Out-V: 12; In V MIN: 18; In V MAX: 36; Out V MIN: 12; Out V MAX: 12; Out-mA: +/-145; Isolation: 8kV; Output: dual; Power Watts or Amps: 3.5W; Package: DIP24; Short Circuit Protec: Yes; MC INPUT-V: 24; MC OUTPUT-V: 12; Length: 31.8 mm; Width: 20.3 mm; Height: 10.2 mm; Packaging Type: Tube Regulated DC/DC Converter; In-V: 4.5-9; Out-V: 12; In V MIN: 4.5; In V MAX: 9; Out V MIN: 12; Out V MAX: 12; Out-mA: +/-145; Isolation: 8kV; Output: dual; Power Watts or Amps: 3.5W; Package: DIP24; Short Circuit Protec: Yes; MC INPUT-V: 5; MC OUTPUT-V: 12; Length: 31.8 mm; Width: 20.3 mm; Height: 10.2 mm; Packaging Type: Tube Regulated DC/DC Converter; In-V: 9-18; Out-V: 12; In V MIN: 9; In V MAX: 18; Out V MIN: 12; Out V MAX: 12; Out-mA: +/-145; Isolation: 8kV; Output: dual; Power Watts or Amps: 3.5W; Package: DIP24; Short Circuit Protec: Yes; MC INPUT-V: 12; MC OUTPUT-V: 12; Length: 31.8 mm; Width: 20.3 mm; Height: 10.2 mm; Packaging Type: Tube Regulated DC/DC Converter; In-V: 36-75; Out-V: 12; In V MIN: 36; In V MAX: 75; Out V MIN: 12; Out V MAX: 12; Out-mA: +/-145; Isolation: 8kV; Output: dual; Power Watts or Amps: 3.5W; Package: DIP24; Short Circuit Protec: Yes; MC INPUT-V: 48; MC OUTPUT-V: 12; Length: 31.8 mm; Width: 20.3 mm; Height: 10.2 mm; Packaging Type: Tube DC-DC Regulated Power Supply Module Regulated DC/DC Converter; In-V: 9-18; Out-V: 15; In V MIN: 9; In V MAX: 18; Out V MIN: 15; Out V MAX: 15; Out-mA: 233; Isolation: 8kV; Output: single; Power Watts or Amps: 3.5W; Package: DIP24; Short Circuit Protec: Yes; MC INPUT-V: 12; MC OUTPUT-V: 15; Length: 31.8 mm; Width: 20.3 mm; Height: 10.2 mm; Packaging Type: Tube
是否Rohs认证 符合 符合 符合 符合 符合 符合
Objectid 7247240106 7247239962 7247240034 7247240178 7247240050 7247240058
Reach Compliance Code compliant compliant compliant compliant compliant compliant
ECCN代码 EAR99 EAR99 EAR99 EAR99 EAR99 EAR99
模拟集成电路 - 其他类型 DC-DC REGULATED POWER SUPPLY MODULE DC-DC REGULATED POWER SUPPLY MODULE DC-DC REGULATED POWER SUPPLY MODULE DC-DC REGULATED POWER SUPPLY MODULE DC-DC REGULATED POWER SUPPLY MODULE DC-DC REGULATED POWER SUPPLY MODULE
是否无铅 不含铅 不含铅 不含铅 不含铅 - 不含铅
