首页 > 器件类别 > 传感器
4525DO-TP5BI015AP
board mount pressure sensors Bd lvl dig out 15psi top Pt i2c 5vex abs
器件类别:传感器
厂商名称:Measurement Specialties(TE Connectivity)
厂商官网:http://www.te.com/usa-en/products/brands/meas.html?tab=pgp-story
4525DO-TP5BI015AP 在线购买
供应商:
器件:4525DO-TP5BI015AP
价格:-
最低购买:-
库存:点击查看
点击购买
MS4525DO
PCB Mounted Digital Output Transducer
Combination Temperature and Pressure
C or SPI Protocol
Differential, Gage, Absolute &
Compound, Vacuum
Temperature Compensated
3.3 or 5.0 Vdc Supply Voltage
Low Power Option Available
(standby < 1uA)
The MS4525DO is a small, ceramic based, PCB mounted pressure transducer from Measurement
Specialties. The transducer is built using Measurement Specialties’ proprietary UltraStable™ process and the
latest CMOS sensor conditioning circuitry to create a low cost, high performance Digital Output Pressure (14bit)
and Temperature (11bit) transducer designed to meet the strictest requirements from OEM customers.
The MS4525DO is fully calibrated and temperature compensated with a total error band (TEB) of less
than 1.0% over the compensated pressure range. The sensor operates from single supply of either 3.3 or
5.0Vdc and requires a single external component for proper operation.
The rugged ceramic transducer is available in side port, top port, and manifold mount and can measure
absolute, gauge, differential, vacuum or compound pressure from 1 to 150psi. The 1/8” barbed pressure ports
mate securely with 3/32” ID tubing.
PSI Pressure Ranges
PCB Mountable
Digital Output
Barbed Pressure Ports
APPLICATIONS
Factory Automation,
Vacuum Switch
Altitude and Airspeed Measurements
Medical Instruments
Leak Detection
STANDARD RANGE (PSI)
Pressure
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP
Absolute
Gauge
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
DS, SS, TP, MM
Differential
DS, SS,TP, MM
DS, SS,TP, MM
DS, SS,TP ,MM
DS, MM
DS, MM
DS, MM
DS, MM
DS, MM
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP
SS, TP, DS
Compound
Vacuum
See Package Configurations: DS= Dual Side Port, SS= Single Side Port, TP= Top Port, MM= Manifold Mount
Only
Protocol is Available on “L” type Pin Styles; Reference Ordering Information for Details
Pin Style “L” is only available SS and MM port types.
MS4525DO
www.meas-spec.com
1/14
October 2013
MS4525DO
BLOCK DIAGRAM
ABSOLUTE MAXIMUM RATINGS
Supply Voltage
Output Current
Storage Temperature
Humidity
Overpressure
Burst Pressure
ESD
Solder Temperature
Conditions
= 25 °C
= 25°C
= 25°C
= 25 °C, both Ports
= 25 °C, Port 1
HBM
Min
2.7
+125
95
Not to Exceed 300
-4
+4
250°C, 5 sec max.
%RH
psi
psi
kV
Non Condensing
See Table 1
EN 61000-4-2
TABLE 1- BURST PRESSURE BY RANGE AND PACKAGE STYLE
Range
001
002
005
015
030
050
DS
90
SS, TP, MM
psi
psi
psi
psi
psi
psi
psi
psi
MS4525DO
www.meas-spec.com
2/14
October 2013
MS4525DO
ENVIRONMENTAL SPECIFICATIONS
Mechanical Shock
Mechanical Vibration
Thermal Shock
Life
MTTF
Conditions
Mil Spec 202F, Method 213B, Condition C, 3 Drops
Mil Spec 202F, Method 214A, Condition 1E, 1Hr Each Axis
100 Cycles over Storage Temperature, 30 minute dwell
1 Million FS Cycles
>10Yrs, 70 ºC, 10 Million Pressure Cycles, 120%FS
Pressure
PERFORMANCE SPECIFICATIONS
Supply Voltage : 5.0V or 3.3 Vdc
Reference Temperature: 25°C (unless otherwise specified)
PARAMETERS
Pressure Accuracy
Total Error Band (TEB)
Temperature Accuracy
Supply Current
Compensated Temperature
Output Pressure Resolution
Output Temperature Resolution
Update Time
Start Time to Data Ready
Weight
Media
-10
-25
MIN
-0.25
-1.0
+105
TYP
MAX
0.25
UNITS
%Span
%Span
ºC
ºC
ºC
bits
bits
mS
mS
grams
NOTES
Non-Corrosive Dry Gases Compatible with Ceramic, Silicon, Pyrex,
PPS, RTV, Gold, Aluminum and Epoxy. See “Wetted Material by Port
Designation” chart below.
1. Proper operation requires an external capacitor placed as shown in Connection Diagram. Output is not ratiometric to supply voltage.
2. The maximum deviation from a best fit straight line (BFSL) fitted to the output measured over the pressure range at 25C. Includes all
errors due to pressure non linearity, hysteresis, and non repeatability.
3. Total pressure error band includes all accuracy errors, thermal errors over the compensated temperature range and span and offset
calibration tolerances. For ideal sensor output with respect to input pressure and temperature, reference Transfer Function charts
below. TEB values are valid only at the calibrated supply voltage.
4. The deviation from a best fit straight line (BFSL) fitted to the output measured over the compensated temperature range.
5. For errors beyond the compensated temperature range, see Extended Temperature Multiplier chart below.
6. Start time to data ready is the time to get valid data after POR (power on reset). The time to get subsequent valid data is then specified
by the update time specification.
7. This product can be configured for custom OEM requirements, contact factory for lower power consumption or higher accuracy.
MS4525DO
www.meas-spec.com
3/14
October 2013
MS4525DO
CONNECTION DIAGRAM
1.
Place 100nF capacitor between Supply and GND to within 2 cm of sensor.
PRESSURE AND TEMPERATURE TRANSFER FUNCTION
Gage, Differential and Compound Pressure Types
3FFF
3CCB
399A
7FF
733
666
Output( 2 HEX counts)
599
4CC
266
199
CC
2000
0333
0000
pe
Ty
pe
Ty
Pmax
Pmin
Pressure
-50 °C
Tmin
-10 °C
85 °C
150 °C
Tmax
A Type:
B Type:
Output( Decimal counts) =
80% *16383
__________
* (Pressure
applied
-Pmin) + 10%*16383
Pmax-Pmin
90%*16383
__________
Pmax-Pmin
* (Pressure
applied
-Pmin) + 5%*16383
Temperature
Output (Decimal Counts)
(Output °C - (-50°C)
Tmin
) * 2047
_________________________________
(150°C
Tmax
- (-50°C)
Tmin
MS4525DO
www.meas-spec.com
4/14
October 2013
MS4525DO
Vacuum Pressure Type
3FFF
3CCB
399A
7FF
733
AT
ype
BT
ype
666
599
4CC
266
199
2000
0333
0000
CC
Pmin(-15 PSI G)
Pressure
Pmax(0 PSI G)
-50 °C
Tmin
-10 °C
85 °C
150 °C
Tmax
A Type:
B Type:
80% *16383
__________
* (Pmax- Pressure
applied
) + 10%*16383
90%*16383
__________
* (Pmax- Pressure
applied
) + 5%*16383
Temperature
Tmin
) * 2047
(150°C
Tmax
- (-50°C)
Tmin
MS4525DO
www.meas-spec.com
5/14
October 2013
参数对比
与4525DO-TP5BI015AP相近的元器件有:4525DO-TP5BI001AP、4525DO-TP5BI005AP、4525DO-TP5BI050AP、4525DO-TP5BI030AP、4525DO-TP5BI100AP。描述及对比如下:
型号 4525DO-TP5BI015AP 4525DO-TP5BI001AP 4525DO-TP5BI005AP 4525DO-TP5BI050AP 4525DO-TP5BI030AP 4525DO-TP5BI100AP
描述 board mount pressure sensors Bd lvl dig out 15psi top Pt i2c 5vex abs board mount pressure sensors Bd lvl dig out 1psi top Pt i2c 5vex abs board mount pressure sensors Bd lvl dig out 5psi top Pt i2c 5vex abs board mount pressure sensors Bd lvl dig out 50psi top Pt i2c 5vex abs board mount pressure sensors Bd lvl dig out 30psi top Pt i2c 5vex abs board mount pressure sensors Bd lvl digout 100psi top Pt i2c 5vex abs
Manufacture - Measurement Specialties Measurement Specialties Measurement Specialties Measurement Specialties Measurement Specialties
- Board Mount Pressure Sensors Board Mount Pressure Sensors Board Mount Pressure Sensors Board Mount Pressure Sensors Board Mount Pressure Sensors
电路设计漫谈之你练到了第几层?
小熊在一些网站上发现了一些比较好的博文,比方说:电路设计漫谈就是一个比较好的博文,笔者收集的电路设...
Fireflye 电源技术
固态电容和电解电容
固态电容全称固态铝质电解电容,它与普通电容(即液态铝质电解电容)最大差别在于采用了不同的介电...
qwqwqw2088 模拟与混合信号
【ST NUCLEO-U5A5ZJ-Q开发板】第4帖 可配置的14Bit的ADC
之前接触的STM32的芯片大部分都是常规的12bit的ADC 这个STM32U5是14...
常见泽1 stm32/stm8
在CCS中如何将.out文件转换为.bin文件
本帖最后由 dontium 于 2015-1-23 12:51 编辑 用hex 在CCS中如何...
anan 模拟与混合信号
[求助]用射极跟随器设计一个视频缓冲器
首先问下什么叫视频缓冲器,它和一般的缓冲器有什么区别或者有什么特殊的要求 谢谢各位大虾了 用射极跟随...
flyaway57 模拟电子
Threadx 事件链 求助
Threadx的事件链(event chaining)这个东西有人用过吗?看了半天没太看明白,求助...
freebsder 嵌入式系统
