Isolated, Wide-Bandwidth Strain Gage Input
5B38
Wide-bandwidth Single-Channel Signal Conditioning
Interfaces, Amplifies, and Filters signals from full-bridge and
half-bridge strain-gage transduces between 300Ω and 10kΩ.
Isolated Precision Output of -5V to +5V.
10 kHz bandwidth of the 5B38 ideally suits to measure
signals that vary rapidly with time.
Module circuitry can withstand 240v rms at the input screw-
terminals.
All 5B38 series modules are mix-and-match and Hot
Swappable.
FUNCTIONAL BLOCK DIAGRAM
Figure 1. 5B38 Functional Block Diagram
APPLICATIONS
Industrial signal conditioning
Industrial signal isolation
Industrial signal filtering
There are also a number of backplanes and mounting sockets
which provide a complete signal conditioning solution for end
users. Each backplane incorporates screw terminals for field
wiring inputs and outputs and cold junction sensors for
thermocouple applications.
These signal conditioners are designed to provide an easy and
convenient solution to signal conditioning problems of both
designers and end users in measurement and control
applications. Typical uses include microcomputer-based
measurement systems, standard data acquisition systems,
programmable controllers, analog recorders and dedicated
control systems. The 5 B series modules are ideally suited to
applications where monitoring and control of temperature,
pressure, flow, rotation and other analog signals are required.
The 5B Series modules and backplanes are approved by Factory
Mutual (FM) and the 5B Series modules are approved by the
Canadian Standards Association (CSA) for use in Class 1,
Division 2, Groups A, B, C, and D locations. These approvals
certify that the 5B Series is suitable for use in locations where a
hazardous concentration of flammable gas may exist only under
fault conditions of operation. Equipment of this category is
called “non-incendive” and they need no special enclosures or
other physical safeguards.
The 5B series modules and backplanes have been tested and
passed the stringent heavy industrial requirements of the
European Union’s electromagnetic compatibility (ENC)
directive – EN50082-1 and EN50081-2. When used according
to installation directions (refer to 5B series User Manual), any
errors caused by EMI/RFI interference will be less than 0.1% of
the full scale 5B measurement range for field strengths up to 10
V/M and frequencies up to 1 GHz.
PRODUCT OVERVIEW
The 5B Series represents an innovative generation of low cost,
high performance plug-in signal conditioners. Designed for
industrial applications, these modules incorporate highly
reliable transformer-based isolation and automated surface-
mount manufacturing technology. They are compact,
economical components whose performance exceeds that
available from more expensive devices. Combining 1500 V rms
continuous isolation, +0.05% calibrated accuracy, small size and
low cost, the 5B Series is an attractive alternative to expensive
signal conditioners and in-house designs
All modules are potted and identical in pin-out and size (2.27” x
2.32” x 0.595”). They can be mixed and matched, permitting
users to address their exact needs, and may be “hot swapped
without disturbing field wiring or power. The isolated input
modules provide 0 to +5V or +5V outputs and accept J, K, T, E,
R, S, N, or B type thermocouples.
These modules feature complete signal conditioning functions
including 240 V rms input protection, filtering, chopper
stabilized low drift +1 uV/oC amplification, 1500 V rms
isolation, and sensor excitation when required.
All modules feature excellent common mode rejection and
meet industrial transient surge withstand specifications.
Rev. 0
Information furnished by Analog Devices is believed to be accurate and reliable.
However, no responsibility is assumed by Analog Devices for its use, nor for any
infringements of patents or other rights of third parties that may result from its use.
Specifications subject to change without notice. No license is granted by implication
or otherwise under any patent or patent rights of Analog Devices. Trademarks and
registered trademarks are the property of their respective companies.
One Technology Way, P.O. Box 9106, Norwood, MA 02062-9106, U.S.A.
www.analog.com
Tel: 781.329.4700
Fax: 781.326.8703
© 2004 Analog Devices, Inc. All rights reserved.
5B38
GENERAL DESCRIPTION
The 5B38 is a wide-bandwidth single-channel signal
conditioning module that interfaces, amplifies, and filters
signals from full-bridge and half-bridge strain-gage transducers
between 300Ω and 10kΩ. The module provides an isolated
bridge excitation of +10V and a protected, isolated precision
output of -5V to +5V. The 10 kHz bandwidth of the module
ideally suits to measure signals that vary rapidly with time, such
as strain on an automobile chassis during a crash test.
The 5B38 protects the computer side from damage due to field-
side over-voltage faults. The module withstands 240V rms at its
input terminals without damage, thereby shielding computer-
side circuitry from field-side over-voltage conditions. In
addition, the 5B38 is mix-and-match and hot-swappable with
all 5B Series modules, so can be inserted or removed from any
socket in the same backplane without disrupting system power.
The 5B38-04 contains bridge completion circuitry, so can
function with half-bridge strain gages. For quarter-bridge
requirements, the user must complete the bridge input to the
half-bridge level externally. The factory can configure the
module for a wide range of input ranges (sensitivities).
A signle-pole anti-aliasing filter resides at each modules input.
A three-pole, low-pass filter in the output stage sets the
bandwidth and yields optimal noise performance for accurate
measurement of small signals in high electrical noise. A
chopper-stabilized input amplifier provides low drift and stable
gain.
Signal isolation by transformer coupling uses a proprietary
modulation technique for linear, stable and reliable
performance. The differential input circuit on the field side is
fully floating, eliminating the need for any input grounding. A
demodulator on the computer side of the signal transformer
recovers the original signal, which is then filtered and buffered
to provide a low-noise, low-impedance output signal. An
additional benefit, the output section acts as a third floating
port, eliminating possible problems from ground loops and
power-supply noise. The output common must be kept within
+3V of power common.
A series output switch eliminates the need for external
multiplexing in many applications. The switch is turned on by
an active-low enable input. If the switch is to be on at all times,
the enable-input should be grounded to power common as it is
on the 5B01 and 5B08 backplanes..
Figure 2
Rev. 0 | Page 2 of 8
5B38
5B38 Models Available
Model
5B38-02
5B38-04
5B38-05
5B38-Custom
Input Bridge Type
Full Bridge
Half Bridge
Full Bridge
Bridge Range
Ωto
10 kΩ
Ωto
10 kΩ
Ωto
10 kΩ
Ωto
10 kΩ
Excitation
+10.0 V
+10.0 V
+10.0 V
+10.0 V
Sensitivity
3 mV/V
3 mV/V
2 mV/V
Output Range
-5 V to +5 V
-5 V to +5 V
-5 V to +5 V
-5 V to +5 V
* Custom Input/sensitivity ranges are available. Refer to configuration guide.
5B38 Specifications
Model 5B38 Full Bridge
Input Ranges
Standard Ranges
Custom Ranges
Output Ranges (R
> 50 kΩ)
±20 mV (2 mV/V Sensitivity)
±30 mV (3 mV/V Sensitivity)
±10 mV to ±500 mV
-5 V to +5 V
Accuracy
Initial @ +25°C
Nonlinearity
Input Offset vs. Temperature
Output Offset vs. Temperature
Gain vs. Temperature
Excitation Voltage Output @ full load
Load Range
Load Regulation
vs. Temperature
Half Bridge Voltage Level
Half Bridge Voltage vs. Temperature
Input Bias Current
±0.08% Span ±10 µV RTI
±0.02% Span
±1 µV/°C
±40 µV/°C
±25 ppm of Reading/°C
+10 V ±3 mV
10 kΩ, minimum; 300
Ω,
maximum
±5 ppm/mA
±15 ppm/°C
N/A
N/A
±3 nA
Input Resistance
Power On
Power Off
Overload
20 MΩ, minimum
40 kΩ, minimum
40 kΩ, minimum
Noise
Input, 0.1 Hz to 10 Hz Bandwidth
Input, 10 kHz Bandwidth
Output, 100 kHz Bandwidth
0.4 µV rms
±70 nV/
Hz
2 µV rms
±250 nV/
Hz
±0.08% Span ±1 mV RTI
+5 V ±1 mV
±15 ppm/°C
Model 5B38 Half Bridge
10 mV peak-peak
Rev. 0 | Page 3 of 8
5B38
Bandwidth, -3 dB
Output Rise Time, 10% to 90% Span
Output Settling Time, to 0.1%
10 kHz
40 µs
250 µs
Common-Mode Voltage (CMV)
Input-to-Output, Continuous
Output-to-Power, Continuous
Transient
1500 V rms, maximum
±3 V, maximum
ANSI/IEEE C37.90.1-1989
Common-Mode Rejection (CMR)
1 kΩ Source Imbalance, 50/60 Hz
Normal Mode Rejection, 50/60 Hz
100 dB
-3 dB @ 10 kHz
Input Protection, Signal and Excitation Voltage
Continuous
Transient
Output Resistance
Voltage Output Protection
Output Selection Time
240 V rms maximum
Continuous Short to Ground
6 µs @ C
load
= 0 to 2,000 pF
Output Enable Control
Max Logic "0"
Min Logic "1"
Max Logic "1"
Input Current "0"
Power Supply Voltage
Power Supply Current
Power Supply Sensitivity
Mechanical Dimensions
+1 V
+2.5 V
+36 V
0.4 mA
+5 V ±5%
200 mA, Full Load; 120 mA, No Load
25 ppm reading/% ±2.5µV RTI/%
2.275" x 2.375" x 0.595"
(57.8 mm x 59.1 mm x 15.1 mm)
Environmental
Temperature Range
Rated Performance
Operating
Storage
Relative Humidity
RFI Susceptibility
-25°C to +85°C
-40°C to +85°C
-40°C to +85°C
0 to 93% @ +40°C non-condensing
±0.5% Span error @ 400 MHz, 5 Watt, 3 ft
7 ms
* Same as full-bridge version.
Includes the combined effects of repeatability, hysteresis, and nonlinearity. Loads heavier than
50 k
will degrade nonlinearity and gain temperature coefficient.
The output common must be kept within ±3 V of power common.
Rev. 0 | Page 4 of 8
5B38
PIN CONFIGURATIONS AND FUNCTIONAL DESCRIPTIONS
x
Figure 3 5B38 Input Field Connections
Table 1. Pin Function Descriptions—
Pin No.
-EXC
LO
HI
+EXC
Figure 4 . Model 5B Series Module, with pin-out assignments.
ESD CAUTION
ESD (electrostatic discharge) sensitive device. Electrostatic charges as high as 4000 V readily accumulate on the human
body and test equipment and can discharge without detection. Although this product features proprietary ESD
protection circuitry, permanent damage may occur on devices subjected to high energy electrostatic discharges.
Therefore, proper ESD precautions are recommended to avoid performance degradation or loss of functionality.
Rev. 0 | Page 5 of 8