首页 > 器件类别 >
MAX9952FCCB+T
Dual Per-Pin Parametric Measurement Units
厂商名称:Maxim(美信半导体)
厂商官网:https://www.maximintegrated.com/en.html
MAX9951/MAX9952
Measurement Units
General Description
The MAX9951/MAX9952 dual parametric measurement
units (PMUs) feature a small package size, wide force and
measurement range, and high accuracy, making the
devices ideal for automatic test equipment (ATE) and other
instrumentation that requires a PMU per pin or per site.
The MAX9951/MAX9952 force or measure voltages in the
-2V to +7V through -7V to +13V ranges, dependent upon
the supply voltage (V
CC
and V
EE
). The devices handle
supply voltages of up to +30V (V
CC
to V
EE
) and a 20V
device-under-test (DUT) voltage swing at full current. The
MAX9951/MAX9952 also force or measure currents up to
±64mA with a lowest full-scale range of ±2µA. Integrated
support circuitry facilitates use of an external buffer ampli-
fier for current ranges greater than ±64mA.
A voltage proportional to the measured output voltage
or current is provided at the MSR_ output. Integrated
comparators, with externally set voltage thresholds,
provide detection for both voltage and current levels.
The MSR_ and comparator outputs can be placed in a
high-impedance state. Separate FORCE and SENSE
connections are short-circuit protected for voltages
from (V
EE
- 0.3V) to (V
CC
+ 0.3V). The FORCE output
also features a low-leakage, high-impedance state.
Integrated voltage clamps limit the force output to lev-
els set externally. The force-current or the measure-cur-
rent voltage can be offset -0.2V to +4.4V (IOS). This
feature allows for the centering of the control or mea-
sured signal within the external DAC or ADC range.
The MAX9951D/MAX9952D feature an integrated 10kΩ
force-sense resistor between FORCE_ and SENSE_.
The MAX9951F/MAX9952F have no internal force-sense
resistor. These devices are available in a 64-pin, 10mm
x 10mm, 0.5mm pitch TQFP package with an exposed
8mm x 8mm die pad on the top (MAX9951) or the bot-
tom (MAX9952) of the package for efficient heat
removal. The exposed pad is internally connected to
EE
. The MAX9951/MAX9952 are specified over the
commercial 0°C to +70°C temperature range.
o
o
o
o
o
o
o
o
o
Force Voltage/Measure Current (FVMI)
Force Current/Measure Voltage (FIMV)
Force Voltage/Measure Voltage (FVMV)
Force Current/Measure Current (FIMI)
Force Nothing/Measure Voltage (FNMV)
Force Nothing/Measure Current (FNMI,
Range E Only)
Termination/Measure Current
Termination/Measure Voltage
Five Programmable Current Ranges
±2µA
±20µA
±200µA
±2mA
±64mA
-2V to +7V Through -7V to +13V Input-Voltage
Range
Force-Current/Measure-Current Adjustable-
Voltage Offset (IOS)
Programmable Voltage Clamps at Force Output
Low-Leakage, High-Impedance Measure, and
Force States
3-Wire Serial Interface
Low 6mA (max) Quiescent Current per PMU
o
o
o
o
o
o
Ordering Information
PART
MAX9951DCCB+D
MAX9951DCCB+TD
MAX9951DCCB-D
MAX9951DCCB-TD
MAX9951FCCB+
MAX9951FCCB+T
MAX9951FCCB-D
MAX9951FCCB-TD
MAX9952DCCB+
MAX9952DCCB+T
MAX9952DCCB-D
MAX9952DCCB-TD
MAX9952FCCB+
MAX9952FCCB+T
MAX9952FCCB-D
TEMP RANGE
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
0°C to +70°C
PIN-PACKAGE
64 TQFP-EPR*
64 TQFP-EPR*
64 TQFP-EPR*
64 TQFP-EPR*
64 TQFP-EPR*
64 TQFP-EPR*
64 TQFP-EPR*
64 TQFP-EPR*
64 TQFP-EP**
64 TQFP-EP**
64 TQFP-EP**
64 TQFP-EP**
64 TQFP-EP**
64 TQFP-EP**
64 TQFP-EP**
Memory Testers
VLSI Testers
System-on-a-Chip Testers
Structural Testers
Pin Configurations and Selector Guide appear at end of
data sheet.
MAX9952FCCB-TD
0°C to +70°C
64 TQFP-EP**
+Denotes
a lead(Pb)-free/RoHS-compliant package.
-Denotes
a package containing lead(Pb).
*EPR
= Top side exposed pad.
D = Dry pack.
**EP
= Exposed pad.
T = Tape and reel.
For pricing, delivery, and ordering information, please contact Maxim Direct
at 1-888-629-4642, or visit Maxim’s website at www.maximintegrated.com.
19-3247; Rev 6; 8/13
MAX9951/MAX9952
Measurement Units
ABSOLUTE MAXIMUM RATINGS
θ
JC
MAX9951_CCB (Note 1) ...........................................+2°C/W
CC
to AGND .......................................................................+20V
EE
to AGND.........................................................................-15V
θ
JA
MAX9952_CCB (Note 1) .........................................+23°C/W
CC
to V
EE
...........................................................................+32V
θ
JC
MAX9952_CCB (Note 1) ...........................................+2°C/W
to AGND............................................................................+6V
Junction Temperature ......................................................+150°C
AGND to DGND.....................................................-0.5V to +0.5V
Storage Temperature Range .............................-65°C to +150°C
Digital Inputs/Outputs ..................................-0.3V to (V
+ 0.3V)
Operating Temperature Range (commercial) ........0°C to +70°C
All Other Pins to AGND ....................(V
EE
- 0.3V) to (V
CC
+ 0.3V)
Lead Temperature (soldering, 10s) .................................+300°C
Continuous Power Dissipation (T
= +70°C)
Soldering Temperature (reflow)
MAX9951_CCB (derate 125mW/°C above +70°C) ...10,000mW
Lead(Pb)-Free Packages.............................................+260°C
MAX9952_CCB (derate 43.5mW/°C above +70°C) .....3478mW
Packages Containing Lead(Pb)......................................+240°C
θ
JA
MAX9951_CCB (Note 1) ...........................................+8°C/W
Note 1:
Package thermal resistances were obtained using the method described in JEDEC specification JESD51-7, using a four-layer
board. For detailed information on package thermal considerations, refer to
www.maximintegrated.com/thermal-tutorial.
Stresses beyond those listed under “Absolute Maximum Ratings” may cause permanent damage to the device. These are stress ratings only, and functional
operation of the device at these or any other conditions beyond those indicated in the operational sections of the specifications is not implied. Exposure to
absolute maximum rating conditions for extended periods may affect device reliability.
DC ELECTRICAL CHARACTERISTICS
(V
CC
= +12V, V
EE
= -7V, V
= +3.3V, T
= +25°C, unless otherwise noted. Specifications at T
= T
MIN
and T
= T
MAX
are guaranteed
by design and characterization. Typical values are at T
= +25°C, unless otherwise noted.) (Note 2)
PARAMETER
FORCE VOLTAGE
Force Input Voltage
Range
Forced Voltage
Input Bias Current
Forced-Voltage Offset
Forced-Voltage-Offset
Temperature Coefficient
Forced-Voltage Gain
Error
Forced-Voltage-Gain
Forced-Voltage Linearity
Error
MEASURE CURRENT
Measure-Current Offset
Measure-Current-Offset
Measure-Current Gain
Error
Measure-Current-Gain
Linearity Error
SYMBOL
IN0_
IN1_
DUT
CONDITIONS
MIN
TYP
MAX
UNITS
EE
+ 2.5
DUT current at full scale
DUT current = 0A
CC
= +12V, V
EE
= -7V
CC
= +18V, V
EE
= -12V
-2
-7
EE
+ 2.5
±1
-25
±100
CC
- 2.5
+7
+13
CC
- 2.5
µA
FOS
+25
mV
µV/°C
FGE
Nominal gain of +1
-1
0.005
±10
+1
ppm/°C
FLER
Gain and offset errors calibrated out (Notes 3, 4)
-0.02
+0.02
%FSR
MOS
(Note 3)
-1
±20
+1
%FSR
ppm/°C
MGE
(Note 5)
-1
±20
+1
ppm/°C
MLER
(Notes 3, 4, 6)
-0.02
+0.02
%FSR
Maxim Integrated
MAX9951/MAX9952
Measurement Units
DC ELECTRICAL CHARACTERISTICS (continued)
(V
CC
= +12V, V
EE
= -7V, V
= +3.3V, T
= T
MIN
and T
= T
MAX
are guaranteed
PARAMETER
Measure-Output-Voltage
Range Over Full-Current
Range
Current-Sense Amp
Offset-Voltage Input
Rejection of Output-
Measure Error Due to
Common-Mode Sense
SYMBOL
IOS
= V
DUTGND
MSR_
IOS
= 4V + V
DUTGND
IOS
Relative to V
DUTGND
+8
+4.4
CONDITIONS
MIN
-4
TYP
MAX
+4
UNITS
CMVR
LER
(Notes 5 and 7)
+0.001
+0.007
%FSR/V
Range E, R_E = 500kΩ
Range D, R_D = 50kΩ
Measure-Current Range
Range C, R_C = 5kΩ
Range B, R_B = 500Ω
Range A, R_A = 15.6Ω
FORCE CURRENT
Input Voltage Range for
Setting Forced Current
Over Full Range
Current-Sense Amp
IOS_ Input Bias Current
Forced-Current Offset
Forced-Current-Offset
Forced-Current Gain
Error
Forced-Current-Gain
Forced-Current Linearity
Error
Rejection of Output Error
Due to Common-Mode
Load Voltage
FLER
(Notes 3, 4, 6)
(Note 5)
(Note 3)
IN0_,
IN1_
IOS
IOS
= V
DUTGND
IOS
= 4V + V
DUTGND
Relative to V
DUTGND
-2
-20
-200
-2
-64
+2
+20
+200
+2
+64
µA
-4
±1
-1
±20
-1
±20
-0.02
+4
+4.4
µA
+1
%FSR
ppm/°C
+1
ppm/°C
+0.02
%FSR
CMRI
OER
(Notes 5 and 7)
Range D, R_D = 50kΩ
-2
-20
-200
-2
-64
+0.001
+0.007
+2
+20
+200
+2
+64
%FSR/V
µA
Forced-Current Range
Range C, R_C = 5kΩ
Range B, R_B = 500Ω
Maxim Integrated
MAX9951/MAX9952
Measurement Units
(V
CC
= +12V, V
EE
= -7V, V
= +3.3V, T
= T
MIN
and T
= T
MAX
are guaranteed
PARAMETER
MEASURE VOLTAGE
Measure-Voltage-Offset
Gain Error
Measure-Voltage-Gain
Measure-Voltage
Linearity Error
Range Over Full DUT
FORCE OUTPUT
Off-State Leakage
Current
Short-Circuit Current
Limit
Force-to-Sense Resistor
SENSE INPUT
Input Voltage Range
Leakage Current
COMPARATOR INPUTS
Input Voltage Range
Offset Voltage
Input Bias Current
VOLTAGE CLAMPS
Input Control Voltage
Clamp Voltage
Accuracy
DIGITAL INPUTS
Input High Voltage
(Note 9)
Input Low Voltage
(Note 9)
Input Current
Input Capacitance
= 5V
IH
= 3.3V
= 2.5V
IL
IN
IN
= 5V or 3.3V
= 2.5V
±1
3.0
+3.5
+2.0
+1.7
+0.8
+0.7
µA
pF
CLLO_
CLHI_
(Note 8)
EE
+ 2.4
-100
CC
- 2.4
+100
mV
EE
+ 2.5
-25
±1
CC
- 2.5
+25
mV
µA
F option only
EE
+ 2.5
-1
CC
- 2.5
+1
nA
LIM-
LIM+
FS
D option only
-1
-92
+65
-65
+92
nA
MLER
(Notes 3, 4, 6)
DUT current = 0A
CC
= +12V, V
EE
= -7V
CC
= +18V, V
EE
= -12V
-0.02
-2
-7
EE
+ 2.5
MGER
Nominal gain of +1
-1
MOS
-25
±100
±0.005
±10
+0.02
+7
+13
CC
- 2.5
+25
mV
µV/°C
ppm/°C
%FSR
SYMBOL
CONDITIONS
MIN
TYP
MAX
UNITS
MSR
Maxim Integrated
MAX9951/MAX9952
Measurement Units
(V
CC
= +12V, V
EE
= -7V, V
= +3.3V, T
= T
MIN
and T
= T
MAX
are guaranteed
PARAMETER
Output High Voltage
Output Low Voltage
High-Impedance-State
Leakage Current
Output Capacitance
DIGITAL OUTPUTS
Output High Voltage
Output Low Voltage
POWER SUPPLY
Positive Supply
Negative Supply
Total Supply Voltage
Logic Supply
Positive Supply Current
Negative Supply Current
Logic Supply Current
Analog Ground Current
Digital Ground Current
Power-Supply Rejection
Ratio
CC
EE
CC
EE
AGND
DGND
PSRR
No load, clamps enabled
No load, all digital inputs at rails
1MHz, measured at force output
60Hz, measured at force output
0.9
1.4
(Note 2)
(Note 2)
+10
-15
+2.375
+12
-7
+18
-5
+30
+5.5
10.0
10.0
1.2
dB
OH
OL
OUT
= 1mA, V
= +2.375V to +5.5V,
relative to DGND
OUT
= -1mA, V
= +2.375V to +5.5V,
relative to DGND
+0.2
SYMBOL
OH
OL
CONDITIONS
= +2.375V to +5.5V, R
PUP
= 1kΩ
PUP
= 1kΩ
±1
6.0
MIN
+0.4
TYP
MAX
UNITS
µA
pF
COMPARATOR OUTPUTS
CC
- V
EE
(Note 10)
Maxim Integrated
参数对比
与MAX9952FCCB+T相近的元器件有:MAX9951、MAX9951FCCB+、MAX9951_09、MAX9952、MAX9952DCCB+T、MAX9952DCCB+、MAX9952FCCB+。描述及对比如下:
型号 MAX9952FCCB+T MAX9951 MAX9951FCCB+ MAX9951_09 MAX9952 MAX9952DCCB+T MAX9952DCCB+ MAX9952FCCB+
描述 Dual Per-Pin Parametric Measurement Units Dual Per-Pin Parametric Measurement Units Dual Per-Pin Parametric Measurement Units Dual Per-Pin Parametric Measurement Units Dual Per-Pin Parametric Measurement Units Dual Per-Pin Parametric Measurement Units Dual Per-Pin Parametric Measurement Units Dual Per-Pin Parametric Measurement Units