Unregulated
Converters
Qualified with 65kV/µs @ Vcommon mode =1KV
EN61010 for test, measurement and lab use
EN60601 for medical applications
Reinforced isolation 6.4kVDC or 8kVDC
Optional continuous short circuit protection
Unique reinforced isolation transformer system
/X2 option for >9mm input/output clearance
DC/DC Converter
RxxPxx/R
The RxxPxxS_D Series of DC/DC Converters are certified to UL/CSA60950-1. This makes them ideal for
safety applications where approved or reinforced isolation is required. The reinforced versions are also
EN61010-1 certified for Lab Equipment Safety.
Selection Guide
Part
Number
RxxP3.3S/R
(3,4,5)
RxxP05S/R
(3,4,5)
RxxP09S/R
(3,4,5)
RxxP12S/R
(3,4,5)
RxxP15S/R
(3,4,5)
RxxP3.3D/R
(3,4,5)
RxxP05D/R
(3,4,5)
1 Watt
SIP7
Single and Dual
Output
nom. Input
[VDC]
5, 12, 15, 24
5, 12, 15, 24
5, 12, 15, 24
5, 12, 15, 24
5, 12, 15, 24
5, 12, 15, 24
5, 12, 15, 24
5, 12, 15, 24
5, 12, 15, 24
5, 12, 15, 24
Output
[VDC]
±3.3
±5
±9
±12
±15
Output
Current
[mA]
303
84
66
±151
±100
±55
±41
±33
Efficiency
typ.
70 - 80
75 - 80
75 - 82
75 - 82
75 - 83
72 - 79
75 - 82
75 - 82
75 - 82
75 - 83
max. Capacitive
Load
(2)
[µF]
2200
470
±1000
±470
±470
±220
±220
RxxP09D/R
(3,4,5)
RxxP12D/R
(3,4,5)
RxxP15D/R
(3,4,5)
Notes:
Note1: Efficiency is tested at nominal input and full load at +25°C ambient
Note2: Max. Capacitive Load is defined as the capacitive load that will allow start up
in under 1 second without damage to the converter
E314885
EN/IEC60950-1 certified
EN/IEC60601-1 certified
UL/CSA 60950-1 certified
ES/CSA60601-1 certified
EN/IEC61010-1 certified
IEC-60601-1 CB report
Model Numbering
R__P__ _/P/X2/R
Input Voltage
Output Voltage
Single
or
Dual
Reinforced Isolation
(3)
Pinout
(4)
Protection
(5)
Notes:
Note3: add suffix „/R6.4“ for 6.4kVDC/1second isolation or „/R8“ for 8kVDC/1second isolation
Note4: add suffix „/X2“ for single output with alternative pinout
Note5: add suffix „/P“ for continous short circuit protection
Ordering Examples:
R05P3.3S/R8/P = 5V Input, 3.3V Output, Single Output, 8kVDC/1s isolation, Continous Short Circuit Protection
R24P05S/R6.4/P/X2 = 24V Input, 5V Output, Single Output, 6.4kVDC/1s isolation, Continous SCP, Alternative Pinout
R12P05D/R8/X2 = ±12V Input, ±5V Output, Dual Output, 8kVDC/1s isolation, Alternative Pinout
www.recom-power.com/eval-ref-boards
www.recom-power.com/bier
www.recom-power.com
REV.: 5/2018
ECO-1
RxxPxx/R
DC/DC Converter
(measured @ Ta= 25°C, nom. Vin, full load and after warm-up unless otherwise stated)
BASIC CHARACTERISTICS
Input Voltage Range
Minimum Load
(5)
Internal Operating Frequency
Output Ripple and Noise
20MHz BW
Notes:
Note5: Operation below 10% load will not harm the converter, but specifications may not be met
20kHz
Condition
Min.
Typ.
±10%
0%
50kHz
85kHz
200mVp-p
Efficiency vs. Load
PxxP05S/R6.4 and PxxP05S/R8
PxxP05D/R6.4 and PxxP05D/R8
Efficiency [%]
R05P05S
R12P05S
R24P05S
Efficiency [%]
R05P05D
R12P05D
R24P05D
Output Load [%]
Output Load [%]
REGULATIONS
Output Accuracy
Line Regulation
Load Regulation
low line to high line, full load
10% to 100% load
3.3V, 5V
9V, 12V, 15V
Condition
Value
±5.0% max.
1.2%/1% of Vin typ.
15% typ.
10% typ.
Tolerance Envelope
Output Voltage [%]
+10%
+5%
Vnom
Typical L
o
ad Line
+5%
-1.5%
-5%
Load [%]
continued on next page
www.recom-power.com
REV.: 5/2018
ECO-2
RxxPxx/R
DC/DC Converter
Deviation vs. Load
PxxP05S/R6.4 and PxxP09S/R8
PxxP05D/R6.4 and PxxP09D/R8
Deviation [%]
-10
-15
R05P05S/R6.4
R12P05S/R6.4
R24P05S/R6.4
Deviation [%]
-10
-15
R05P05D/R6.4
R12P05D/R6.4
R24P05D/R6.4
Output Load [%]
Output Load [%]
PROTECTIONS
Short Circuit Protection (SCP)
without Suffix “/P”
with Suffix “/P”
tested for 1 second
/R6.4
/R8
/R6.4
/R8
Value
1 second
continuous
6.4kVDC
8kVDC
3.2kVAC/60Hz
4kVAC/60Hz
15GΩ min.
Isolation Voltage
(8)
Isolation Resistance
Isolation Capacitance
Leakage Current
Insulation Grade
Means of Protection
Internal
External
I/P to O/P
rated for 1 minute
4.0pF min. / 10pF max.
<0.01µA max.
reinforced
34Vrms
2MOPP
>4.8mm
>4.8mm
clearance/creepage
clearance/creepage
Notes:
Note7: For repeat Hi-Pot testing, reduce the time and/or the test voltage
ENVIRONMENTAL
Maximum Case Temperature
Operating Altitude
Operating Humidity
Pollution Degree
MTBF
according to MIL-HDBK-217F, G.B.
www.recom-power.com
REV.: 5/2018
ECO-3
Condition
without derating @ free air convection (see graph)
Value
-40°C to +85°C
+105°C
3000m
non-condensing
+25°C
+85°C
95% RH max.
PD2
2974 x 10
hours
728 x 10
hours
RxxPxx/R
DC/DC Converter
Derating Graph
(@ Chamber and free air convection)
Output Load [%]
-40 -30 -20 -10 0 10 20 30 40 50 60 70 80 90 100 110
85
Ambient Temperature [°C]
SAFETY AND CERTIFICATIONS
Certificate Type (Safety)
Information Technology Equipment, General Requirements for Safety
Medical Electric Equipment, General Requirements for Safety and
Essential Performance
Essential Performance. (CB Scheme)
Safety requirements for electrical equipement for measurement, control
and laboratory use
EAC
RoHs 2+
Report / File Number
1605077-14
2207629
2207629
E314885-A5-UL
E314885-A5-CB-1
1205098-1
T1301251-313
RU-AT.37.02367
Standard
EN60950-1: 2006 + AM2:2013
IEC60950-1-2005 , 2nd Edition + AM2:2013
ANSI/UL60950-1, 1st Edition
CAN/CSA C22.2 No. 60950-1
UL60601-1, 1st Edition
CAN/CSA C22.2 No. 60.1-M90
ANSI/AAMI ES60601-1:2005 +A2:10
CAN/CSA-C22.2 No. 60601-1:2008
IEC60601-1:2005 + C2:2007
EN60601-1:2006
EN61010-1:2010
IEC61010-1:2010, 3rd Edition
TP TC 004/2011
RoHS (10/10)
EMC Compliance
Electromagnetic compatibility of multimedia equipment -
Emission requirements
Condition
with external filter
Standard / Criterion
EN55032, Class A/B
www.recom-power.com
REV.: 5/2018
ECO-4
RxxPxx/R
DC/DC Converter
EMC Filtering Suggestions according to EN55032 Class A and Class B
Class A
+V
IN IN
+V
C1 C1
-V
IN IN
-V
-V
OUT OUT
-V
-V
IN IN
-V
C2 C2
+V
OUT OUT
+V
Class B
L1 L1
+V
IN IN
+V
C1 C1
C2 C2
-V
OUT OUT
-V
+V
OUT OUT
+V
C3 C3
Component List Class A
Model
RxxPxx/R6.4
RxxPxx/R8
Component List Class B
C2
2n2F 8kV
Vishay HGZ222MBP
2n5F 10kV
Vishay HGZ222MBP
C1
10µF
10µF
Model
RxxPxx/R6.4
RxxPxx/R8
C1
10µF
10µF
L1
470µH
WE 7447471471
470µH
WE 7447471471
C2
10µF
10µF
C3
2n2F 8kV
Vishay HGZ222MBP
2n5F 10kV
Vishay HGZ222MBP
DIMENSION AND PHYSICAL CHARACTERISTICS
Material
Dimension (LxWxH)
Weight
case
potting
PCB
Value
non-conductive black plastic, (UL94 V-0)
silicon rubber compound, (UL94 V-0)
FR4, (UL94 V-0)
19.5 x 9.8 x 12.5mm
4.3g typ.
Dimension Drawing (mm)
embossed logo
19.5
9.8
Marking
Pin Connection
Pin # Single
+Vin
-Vin
-Vout
No Pin
+Vout
Tolerance:
Dual
+Vin
-Vin
-Vout
Com
+Vout
/X2
+Vin
-Vin
No Pin
-Vout
+Vout
0.51
12.5
xx.x= ±0.5mm
xx.xx= ±0.25mm
4.10
0.50
+0.10/-0.05
0.25
±0.05
Recommended Footprint Details
Bottom View
1 2
2.54
2.4
1 2
5 6 7
Top View
3x2.54= 7.62
1.9
6x2.54= 15.24
2.54
www.recom-power.com
REV.: 5/2018
ECO-5