首页 > 器件类别 >
MPU400MS24YZI
MPU-400M Series
厂商名称:MicroPower Direct ( MPD )
厂商官网:http://www.micropowerdirect.com
MPU-400M
Single Output, 400W
Medical Approved
AC/DC Power Supplies
Electrical Specifications
Key Features:
Compact 400W Supply
EN 60601-1 3
RD
Edition
PFC to EN 61000-3-2 “D”
Low Profile <1U Height
700W Peak Power
Universal 90-264 VAC Input
Current Share Option
Specifications typical @ +25°C, nominal input voltage & rated output current, unless otherwise noted. Specifications subject to change without notice.
Input
Input Voltage Range
Input Frequency
Input Current, Full Load
Inrush Current, Cold Start
Leakage Current
Power Factor Correction
Input Protection
Input Undervoltage Protection
Conditions
Universal
90 VAC
230 VAC
Min.
90
47
Typ.
264
63
Units
VAC
Hz
µA
6.35
Meets EN 61000-3-2 Class D
Two T8A/250V Fuse
Under 80 VAC ±5 VAC Unit Shuts Down; Unit Recovers Over 86 VAC
Conditions
By Trim Pot
Min.
Typ.
±5.0
±1.0
Units
mSec
ºC
%/ºC
mS
Output
Output Voltage Adjustment
Output Regulation, See Note 1
Hold Time
Ripple & Noise (20 MHz), See Note 2
Overload Protection
Over Voltage Protection
Over Temperature Protection
Temperature Coefficient
Transient Recovery Time, See Note 4
Transient Response Deviation
Overshoot/Undershoot
Turn On Delay
Output Short Circuit
120 VAC, 80% Load
See Model Selection Guide
Foldback Circuit, Autorecovery
110
140
>130% of Rated Output Voltage. Recycle AC Input.
Autorecovery
+85
±0.04
50% Load Change
5.0
At Turn On/Off
±5.0
230 VAC
Continuous With Autorecovery
Conditions
Input - Output
Input - FG (Frame Ground)
Primary - Core
Fixed
Min.
4,000
1,500
1,500
Typ.
General
Isolation Voltage, See Note 5
Switching Frequency
Units
VAC
kHz
Interface Signals
LED Power Supply On
AC Fail (Option)
Current Monitor
Current Share (Option)
Margin (Option)
Remote Sense
Bi-color LED Is Green For Power On; Orange When Protection Enabled
Signal goes low (0V) if input drops below 80 VAC (± 5VAC). Signal goes
high (5V) if input rises above 86 VAC
A 0.5 to 3.0V Output That Represents 0% to 100% Output Current
For Sharing Up To Four Units. Contact The Factory For Details
For remote output voltage adjustment. Contact The Factory For Details
Compensates For Up To A 0.5V Line Drop
Goes TTL high 100 to 500 mS after regulation. Goes low at least 1 mS
before the loss of regulation.
A TTL low signal inhibits the output.
Conditions
Min.
Typ.
Units
+70
ºC
2.5%/ ºC from +50 ºC to + 70 ºC
-20
+85
ºC
RH, Non-condensing
90
Ambient
MicroPower Direct
292 Page Street
Suite D
Stoughton, MA 02072
USA
T:
(781) 344-8226
F:
(781) 344-8481
E:
sales@micropowerdirect.com
W:
Power Good Signal
Remote On/Off
Environmental
Output Derating
Storage Temperature Range
Cooling
Operating Humidity
Reliability Specifications
MTBF
Safety Standards
EMI Compliance
EMS Immunity Compliance
Vibration
Conditions
Min.
Typ.
Units
MIL HDBK 217F, 30ºC, Gnd Benign
kHours
EN 60601 3
RD
Edition (UL)
Compliance to EN 55022 (CISPR22) Class B; EN 61000-3-2, 3
EN 61000-4-2,3,4,5,6, 8,11; EN 55024, CE Marked (LVD)
Sinusoidal 5~50 Hz, Acceleration ±7.35 m/s
on X, Y & Z Axis
Model
Number
MPU-400MS-12
YZI
MPU-400MS-24
YZI
MPU-400MS-48
YZI
Output Voltage
(VDC)
PreSet
Range
Output Current
(A)
Convection
20.83
10.41
5.20
With
23 CFM
33.33
16.66
8.33
Notes:
1. Output regulation includes line & load.
Ripple
& Noise Efficiency
2. Ripple & noise is measured from 10 Hz to 20 MHz. Connection to the unit is made
with a 0.1 µF ceramic capacitor & a 22 µF electrolytic capacitor connected in parallel.
(% p-p)
3. A 1% minimum load is required to maintain regulation & ripple specifications.
4. Transient recovery is measured to within a 1% error band for a load step change
of 50% to 100%.
5. Isolation specifications are production HI-Pot tested for 3 seconds.
6. The full output range (see table) is covered in the safety agency certification. Standard
models are factory set to the Preset voltage, but may be set to other levels within
the range without affecting the agency certification. For more information, contact
the factory.
7. Output power is given for the factory preset voltage. The maximum continuous
output power level is 400W with 23 CFM airflow. All models provide a peak power
level of 700W for a maximum duration of 500
µS.
the factory.
8. Each unit includes two (250V/8A) fuses. Since these fuses are not field replaceable,
it is recommended that an external fuse of the same size be used on the input of
the power supply for protection.
12 VDC 12.0 - 15.0
24 VDC 22.0 - 30.0
48 VDC 42.0 - 58.0
±1%
±1%
±1%
80%
80%
80%
Model Number
Models with other output voltage levels
are available (i.e. 5 VDC, 36 VDC, etc)
Contact the factory for details at:
Mechanical Dimensions
Connections
Input Connector (CN1):
Terminal Block:
Howder HD-121-3P: M3.5 Screws
3 pins, 9.5 mm Centers
Molex Mating Part No:
Molex 09-91-0700 or equivalent
(Seven pins, Five pins used)
Howder
Pin
Function
Pin
Molex
Function
AC-Line 1,2 AC-Line
2 AC-Neutral 4,5 AC-Neutral
3 Field Ground 7 Field Ground
Output Connector (CN2):
Terminal Block:
Howder HD-121-6P: M3.5 Screws
6 pins, 9.5 mm Centers
Molex 09-91-1600 or equivalent
Howder
Pin
Function
Pin
Molex
Function
1-3
4-6
-V
OUT
+V
OUT
1-8
9-16
-V
OUT
+V
OUT
Logic Signal Connector (CN3):
Mating Part No:
JST XHP-9 or equivalent
(CHYAO SHIUNN JS-2001-09)
U-Chassis Cover
(C Suffix)
Pin
Function
+ Remote Sense
- Remote Sense
Remote On/Off
Power Good
AC Fail
Common
Current Share
Current Monitor
Margin
Fan Driver Connector (CN4):
Mating Part No:
JST XHP-2 or equivalent
(CHYAO SHIUNN JS-2001-02)
Pin
Function
Plus
Minus
MicroPower Direct
We Power Your Success - For Less!
292 Page Street Ste D Stoughton, MA 02072
Notes:
All dimensions are typical in inches (mm)
Tolerance x.xx =
0.02 (
0.50)
FAX: (781) 344-8481
TEL: (781) 344-8226
E-Mail: sales@micropowerdirect.com
参数对比
与MPU400MS24YZI相近的元器件有:MPU-400M、MPU400MS12YZI、MPU400MS48YZI。描述及对比如下:
型号 MPU400MS24YZI MPU-400M MPU400MS12YZI MPU400MS48YZI
描述 MPU-400M Series MPU-400M Series MPU-400M Series MPU-400M Series
请问,CS+ for CACX 开发环境,现在免费使用吗
请问,CS+ for CACX 开发环境,现在免费使用吗 请问,CS+ for CACX 开发环...
wdliming 瑞萨电子MCU
AD铺铜的疑问
这种铺铜方式有什么优点,发现用的不多。 AD铺铜的疑问 是不常用, 相同网络的焊盘、覆铜进行...
朴树先生 PCB设计
提问该分版斑竹,关于该分版
怎么有了这个版块,这个斑竹跑哪里去了? 我正在用这个公司的板子,需要一些问题的了解, 既然开版了,...
zqzq501311 红色飓风FPGA专区
lm3m8962上SD卡文件系统读写例程串口助手显示乱码
在lm3m8962上跑SD卡文件系统读写程序例程,为什么显示的是乱码? lm3m8962上SD卡文...
nillq 微控制器 MCU
《深度强化学习实战》第一章学习及gym代码验证
深度强化学习是机器学习的一个子领域,它将深度学习模型应用于强化学习任务中。书中以图像...
lospring 嵌入式系统
虚拟机同时安装xp和win7的问题
我虚拟机同时安装xp和win7,xp可以和电脑互传数据,win7不可以,不知道什么原因,设置了共享...
chenbingjy WindowsCE
